This Information Applies To: Agilent 6000 Series LC/MS Systems equipped with MS40+ rotary vane pumps
Issue
Replacing the MS-40+ (rough pump) fluid periodically is a key maintenance procedure to improve vacuum performance and minimize the possibilities of pump failure.
Background
The rough pump fluid should be replaced if the pump fluid appears dark or cloudy or after every six months of operation.
Items Required
Caution: Use only the rough pump fluid appropriate for your pump (AVF 60 Gold oil for Agilent MS40+). The use of any other fluids can substantially reduce pump life and will in validate the pump warranty.
|
Steps to Follow
Warning: Do not touch the fill cap or rough pump until they are cool. The fill cap and pump can be hot. Wear chemical resistant gloves and safety glasses when replacing the pump fluid.
Never add or replace the rough pump fluid while the pump is on. Hot oil can splash out and cause injury.
Do not touch the fluid. The residue from some samples can be toxic. Properly dispose of the used fluid as per the AVF 60 Gold fluid MSDS shipped with the product.
|
Tip: Draining the pump fluid while it is still warm, but not hot, will help with removal of the old fluid from the MS40+ pump.
|
Caution: The older petroleum-based pump oils SW60 (supplied in kit p/n 5190–9008) or AVF60M (p/n 6040–1361) should not be mixed with the current synthetic pump AVF 60 Gold fluid. There is no need to rinse out the old oil, but allow sufficient time for the pump to fully drain.
|
Figure 3. Refilling pump fluid to the maximum level marked next to the fluid level window