Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles How to Remove and Reinstall the Needle in Agilent G4226X/G1367E High Performance (HiP) Autosamplers
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  6 Feb 2023
  • Modified :  22 Jun 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Injectors and Sample Thermostats
  • Product Name :  1260 Infinity LC System 1290 Infinity LC System
  • Task :  Maintenance
Related Articles
  • Removing and Reinstalling the Needle Seat in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

    This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler Issue Replacing the Needle Seat of the Autosampler is required when: The needle seat is visibly damaged. Leaks or blockages in the Autosampler are observed from the needle seat. Replacement the needle seat is performed as part of yearly maintenance. Tool Required Wrench, open end 1/4 - 5/16 inch (p/n 8710-0510 ) Parts Required For an Agilent 1290 Infinity (G4226A...
  • How to Change the Needle on a G7167x Multisampler

    This Information Applies To: Agilent G7167A/B Multisampler. Issue: Required when: The needle is visibly damaged. Leaks or blockages in the vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of yearly maintenance Steps to follow: Tools Required: 8710–0510 - Open-end wrench 1/4 - 5/16 inch Parts Required: G4267-87201 - Needle assembly OR G4267-87210 -...
  • How to Change the Needle Seat on a G7167x Multisampler

    This Information Applies To: Agilent G7167x Multisampler modules Issue: Required when: The needle seat is visibly damaged, blocked, or is leaking. Leaks or blockages in the Vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of the yearly maintenance procedures. Steps to follow Tools Required: 8710–0510 - Open-end wrench 1/4 - 5/16 inch Flathead screwdriver...
  • How to Change the Needle and Needle Seat in an Agilent G7129x Vialsampler

    This Information Applies To: Agilent G7129A/B/C Vialsampler Issue: Replacing the needle and needle seat of the Vialsampler is required when: The needle is visibly damaged. Leaks or blockages in the vialsampler are observed, especially between the needle and needle seat. The limit for the needle interaction EMF counter is exceeded. The needle needs to be replaced as part of yearly maintenance. Over time, deposits can block the needle and the needle seat causing pressure issues...
  • Troubleshooting "Metering Home Failed" Error in Agilent Autosamplers

    This Information Applies To: Agilent Autosamplers G1313A, G1329A/B, G4226A, G1367A/B/C/D/E, G7129A/B/C. Issue: If during initialization or during a sequence, the "Metering Home Failed" error occurs, this article will guide through the procedure of identifying the cause. Sometimes, this error may also appear as "Cannot Find Metering Sensor" . Steps to follow: Often the reason is a blockage in the injection system. Remove the capillaries into and out of the metering device ( Figure 1...

How to Remove and Reinstall the Needle in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

Answer

This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers


Issue

Replacing the Needle Assembly of the Autosampler is required when:

  • The needle is visibly damaged.
  • Leaks or blockages in the Autosampler are observed from the needle.
  • The limit for the needle interaction EMF counter is exceeded.
  • Replace the needle as part of yearly maintenance.

Tool Required

  • Wrench, open 1/4 - 5/16 inch (p/n 8710-0510)

Parts Required

  • Needle assembly (p/n G4226-87201)

See also: Needle & Needle Seats for HPLC

Steps to Follow

  
Warning: This procedure risks exposure to hazardous, toxic, or flammable solvents and reagents. Wear appropriate PPE as described in the material handling and safety data sheet supplied by the chemical vendor, and always follow good laboratory practice.

 

   Warning: Risk of injury by uncovered needle. 
  • Be careful working at the needle carrier assembly and appropriate PPE when removing the needle assembly.
  • Use the silicon safety tube supplied with every new needle.

 

  
Caution: It is recommended to always exchange the needle assembly and the needle seat at the same time to prevent premature leakage. See Article: How to remove and reinstall the Needle Seat in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

Figure 1 shows main assemblies of an Agilent G4226X/G1367E High Performance (HiP) Autosamplers.

Figure 1. Main Assemblies of an Agilent G4226X/G1367E High Performance (HiP) Autosamplers.
1. Needle Assembly, 2. Needle Carrier, 3. Transport Assembly, 4. Blocked needle seat, 5. Loop capillary, 6. Analytical Head, 7. Leak parts, 8. Needle seat

To remove the Needle Assembly.

1. Finish any pending or running acquisition job.

2. Stop the flow at the pump and remove the solvent lines from the eluent bottles to avoid spilling solvent.

3. Close the shutoff valves at the pump if available on your pump module.

4. In the Agilent Lab Advisor Software, connect to your LC system then click Service and Diagnostics> Autosampler> Maintenance Positions> Change Needle/Seat. Instructions on downloading and installing Lab Advisor on your Workstation PC are provided in the article How to Install Lab Advisor.

5. Click Start to start the maintenance task. Confirm the part number for the needle listed in the pop-up configuration window matches the new needle being installed in the autosampler (Figure 2).



Figure 2. Run the change needle/seat maintenance task in Lab Advisor.

6. Open the front door (1), remove the side door (2) (Figure 3). Turn the Needle carrier 90 ° clockwise. (Figure 4)



Figure 3. Open the front door and remove the side door.
1. Front door, 2. Side door



Figure 4. Turn the Needle carrier 90 ° clockwise.
.

7. Flip the Leak guide open then pinch the holder clamp and pull back to remove the needle from the arm. (Figure 5)



Figure 5. Open the leak guide and remove the needle from the arm.

8. Attach a 5/16-inch wrench to the slot in the body of the needle assembly and use a 1/4-inch wrench to loosen the fitting of the loop capillary, then remove the needle assembly. (Figure 6)



Figure 6. Loosen the fitting of the loop capillary and remove the needle assembly.

To install the Needle Assembly.

9. Push the silicon safety tube delivered with every needle over the needle to prevent injury.

10. Insert the loop capillary into the needle assembly and tighten the fitting hand tight. Attach a 5/16-inch wrench to hold the position at the needle assembly and use a 1/4-inch wrench to tighten the fitting of the loop capillary. (Figure 7)



Figure 7. Insert the loop capillary into the needle assembly and tighten the fitting of the loop capillary.

11. Pinch the holder clamp and re-insert the needle assembly into the needle carrier then close the leak guide. (Figure 8)



Figure 8. Re-insert the needle assembly into the needle carrier and close the leak guide.

12. Check the alignment of the needle in the needle pusher of the needle carrier by viewing from several directions to see that it is aligned in the center of the needle pusher. (Figure 9)

  
Caution: The needle must be centered in the needle pusher as all alignment by the Autosampler is calculated from the needle pusher position.



Figure 9. Check the alignment of the needle in the needle pusher.

13. Remove the silicon safety tube from the needle then reinstall the side door and close the front door.

14. Go back to Lab Advisor and click End to finish the Maintenance Task. Click System Overview and then click Disconnect to disconnect your instrument from Lab Advisor. (Figure 10)



Figure 10. Close the Change Needle/Seat Maintenance task and disconnect the instrument from the Lab Advisor.

  
Tips: If possible, run a system pressure test with the port No. 6 of the injection valve blanked using the appropriate blanking nut to ensure the autosampler flow path is pressure tight to a pressure higher than your application requires.
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help