Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Hello Samehalsharawy ,
Can you please elaborate on the statement we uninstalled the cylinder expecting the flow to come down to zero but it didn't. It stayed at 50?
If no physical tank is connected to the GC, the GC gas connections are capped…
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
ICP-MS Instrument Control or Data Analysis Software will not open, or crashes shortly after opening.
Background
MassHunter uses configuration files to store...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.4 and above (Rev C. and Rev D.)
Issue
The Glitch Resetter resets configuration setting files for MassHunter and can be used when the MassHunter software is exhibiting strange behavior...
This Information Applies To: Agilent ICP-MS MassHunter Software 5.x (Rev D)
Issue
It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x (Rev C)
Issue
It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading...
This Information Applies To: Agilent ICP-MS MassHunter Software revisions 4.x and 5.x
Issue
How to clear or archive the acquisition queue? How to resolve queue task file errors?
Background
The Acquisition Queue ( Figure 1 ) lists all tasks...
This Information Applies To: Agilent Fluorescence Detectors G1321A, G1321B, G7121A, G7121B
Issue
This article describes how to diagnose a leak in your fluorescence detector (FLD) module.
Steps to Follow
Confirm the leak error originated in...
This Information Applies To: Agilent ICP-MS MassHunter 5.1 Software, SPS 4 Autosampler
Issue
The SPS-4 autosampler (ALS) may need to be reconfigured after the Glitch Resetter tool is used to reset the software or after a change from a different...
This Information Applies To: Agilent G4208A Instant Pilot
Issue
This article describes how to switch the illumination of the autosampler on or off (if available) using the G4208A Instant Pilot.
Items Required
G4208A Instant Pilot for Agilent...
This Information Applies To: Agilent 7683B Automatic Liquid Sampler
Issue:
O nly use this alignment procedure if the injector hardware configuration has changed or if the Align Mode light comes on.
Agilent does not recommend performing this procedure...
This Information Applies To: Agilent MassHunter data acquisition software and MassHunter Quantitative Data Analysis Software
Issue
This article explains the difference between quantifier and qualifier ions and provides some general guidance on quantifier...