This Information Applies To: Agilent Fluorescence Detectors G1321A, G1321B, G7121A, G7121B
Issue
This article describes how to diagnose a leak in your fluorescence detector (FLD) module.
Steps to Follow
Confirm the leak error originated in the FLD module before starting this procedure. See Determining which Module is Causing the Shutdown Error Due to a Leak
Warning: The following procedure risks exposure to solvents in the flow path of the FLD flow cell. Always wear safety goggles and appropriate gloves for the solvents used in your LC system before performing the following steps.
|
Figure 1. Outlet and Inlet capillary locations (blue arrows)
Caution: Confirm that the FLD lamp is tuned off before removing the flow cell.
|
For replacement flow cell information and pricing, see the FLD Supplies for HPLC page in Agilent.com, or contact your local Agilent Service Representative for further details.
Learn how to effectively maintain or operate your Agilent LC:
Agilent InfinityLab LC Series 1260 Infinity II Fluorescence Detectors User Manual on Agilent.com Agilent 1260 Infinity Fluorescence Detector User Manual on Agilent.com |