Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-MS Portal
  • Agilent Knowledge Portal
ICP-MS Portal
Articles ICP-MS MassHunter 4.x Software Settings to Note
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-MS Portal Home
  • ICP-MS Portal Articles
  • ICP-MS Portal Videos
  • More
  • Cancel
  • Created :  19 Apr 2023
  • Modified :  8 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP-MS Software
  • Product Name :  ICP-MS MassHunter Software
  • Task :  Best Practices
Related Articles
  • ICP-MS MassHunter 5.x Software Settings to Note

    This Information Applies To: Agilent ICP-MS MassHunter Software 5.x (Rev D) Issue It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading/reloading MassHunter software. Caution : Maintaining software settings is the responsibility of the user. This guide does not show every possible MassHunter software settings window, and there is no way for Agilent to know what settings...
  • ICP-MS MassHunter Software Will Not Open or Crashes After Opening

    This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x Issue ICP-MS Instrument Control or Data Analysis Software will not open, or crashes shortly after opening. Background MassHunter uses configuration files to store software settings and information. Sometimes these files can become corrupt, preventing software operation. Before running the Glitch Resetter (which is a program used to reset all configuration files), it is possible to rename individual files manually...
  • How to Run ICP-MS MassHunter Glitch Resetter

    This Information Applies To: Agilent ICP-MS MassHunter Software 4.4 and above (Rev C. and Rev D.) Issue The Glitch Resetter resets configuration setting files for MassHunter and can be used when the MassHunter software is exhibiting strange behavior such as: Wrong arrangement of user interface or missing icon/menu. MassHunter Instrument control or Data Analysis crashes or will not open. Unable to complete startup because hardware settings are too far from the correct value. Connection...
  • How To Configure the SPS 4 Autosampler on the Agilent ICP-MS MassHunter Software?

    This Information Applies To: Agilent ICP-MS MassHunter 5.1 Software, SPS 4 Autosampler Issue The SPS-4 autosampler (ALS) may need to be reconfigured after the Glitch Resetter tool is used to reset the software or after a change from a different autosampler, such as the ASX-520. This process will aid in configuring the SPS 4 autosampler inside the ICP-MS MassHunter software (in this article ICP-MS MassHunter version 5.1 was used). Steps to Follow In the Home tab, ensure that the Use...
  • Elevated Doubly Charged Ratio in ICP-MS

    This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900) Issue Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an issue with the sample introduction system. Background The Doubly Charged Ratio (%) is defined as: mass 70 (Ce++) / mass 140 (Ce) x 100 Zn, Ge, ClCl, GaH, and FeO are some examples of contaminants/interferences that contribute to elevated counts...

ICP-MS MassHunter 4.x Software Settings to Note

Answer

This Information Applies To: Agilent ICP-MS MassHunter Software 4.x (Rev C)


Issue

It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading/reloading the ICP-MS MassHunter software.


 

  
Caution: Maintaining software settings is the responsibility of the user. This guide does not show every possible MassHunter software settings window, and there is no way for Agilent to know what settings a user had previously.

 

Steps to Follow

This article will cover how to capture the following software settings:

  1. Communication
  2. MassHunter Configuration
  3. Lens Parameters for Semi Auto Tune
  4. Sample Introduction
  5. Plasma Ignition Mode
  6. Ignition Sequence
  7. User Tune (if applicable)
  8. Startup Tasks

 

  1. Communication (Figure 1)
    1. Click the Hardware icon, then the Mainframe image and from the context menu, click Communication.
    2. Capture the instrument IP address and Host Name.
    3. Click Cancel.



      Figure 1. Communication Settings

  2. MassHunter Configuration (Figure 2)
    1. From the Tools menu, click MassHunter Configuration.
      Note: if the menu bar is not visible, press F10 on the keyboard.
    2. Capture the Optional Functions, Data Options, and Report Option tabs.
    3. Click Cancel.



      Figure 2. MassHunter Configuration Settings

  3. Lens Parameters for Semi Auto Tune (Figure 3)
    1. From the menu bar, click Tools >Tune > Select Lens Parameters for Semi Auto Tune...
      Note: if the menu bar is not visible, press F10 on the keyboard.
    2. Capture the selected parameters.
    3. Click Cancel.



      Figure 3. Lens Parameters for Semi Auto Tune Settings

  4. Sample Introduction (Figure 4)
    1. Click the Settings icon.
    2. Click the arrow on the left of Hardware.
    3. On the left pane, click Sample Introduction.
    4. Capture the Sample Introduction Properties page.
    5. On the right of the Autosampler section, click Edit.
    6. Capture the settings of each tab of the Configure Autosampler dialog box.
    7. Click Cancel.
    8. Click Close.



      Figure 4. Sample Introduction Settings

  5. Plasma Ignition Mode (Figure 5)
    1. Click the Settings icon.
    2. Click the arrow on the left of Hardware.
    3. On the left pane, click Plasma.
    4. Capture the Plasma Ignition Mode setting.
    5. Click Close.



      Figure 5. Plasma Ignition Mode Settings

  6. Ignition Sequence (Figure 6)
    1. On the right of the Plasma Icon, click the drop-down arrow, and select Configure Ignition Sequence.
    2. Capture the ignition sequence settings.
    3. Click Cancel.



      Figure 6. Ignition Sequence Settings

  7. User Tune - only applicable if using User Tune for Startup, or Global Tune. If you do not use User Tune, continue to step 8. 
    1. Check if you use User Tune (Figure 7). If User Tune is selected for Configure Ignition Sequence (from step 6), or if you see a globe on the Tune icon, you use User Tune. 



      Figure 7. Check if User Tune is needed. If either of these options are selected, you use User Tune.

    2. Click the Tune icon
    3. In the Plasma section, select the All Parameters check box.
    4. Capture the User Tune settings (Figure 8). 



      Figure 8. User Tune settings


  8. Startup Tasks
    1. Click the Plasma icon.
    2. In the Startup section, capture the Hardware Settings table and the Set Peripump table.
      • If Select Custom Settings is disabled, click Set PeriPump/Isis under Standard Settings and capture the table (Figure 9).



        Figure 9. Startup Settings without Custom Settings Enabled 

      • If Select Custom Settings is enabled (the button is orange), click the Set button on the Torch Axis line and capture the table (Figure 10). These settings will need to be inputted for each line when reconfiguring.



        Figure 10. Startup Settings with Custom Settings Enabled 

  9. After running the Glitch Resetter, renaming configuration settings files or upgrading/reloading the software, enter the settings captured in steps 1 to 8.

 

  Learn how to effectively maintain your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter
Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent education
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help