We've been having issues with our LC-MS/MS 1260/6490 and 1290/6470 systems giving "LC Offline" errors when the instrument is all ready to go and we start a worklist. It clearly is connected, because it's showing the LC status the entire time. We're using…
From a GCMS perspective, 1.9 ml/min into the ion source is not optimal. The MS prefers down around 1.2 ml/min. Sensitivity is better and spectral tilt is more typical.
Septum head or septumless head? What is your gas supply pressure? When was the last…
Hello,
Question 1: Are you opening a snapshot or using review completed injections to open the in-progress injections? Is it possible you were on a replicate injection when trying to change the type.
Question 2: Andy is correct you must have the…
Helllo mighty Marty,
we have only one Infinity II device, so it can be useful. It works same as your example. But if we are going to compare this system with GC catalog it seems to be very painful to write all the info again and again when you are going…
Hello payninater ,
To utilize the inherited calibration curve you will need to copy the cal levels to the other targets. If you are doing this make certain your unknown targets have the same ISTD assigned. Quant does not verify this and simply applies…
Hello payninater ,
It is not possible to enter responses manually in the Method Editor.
As far as I know there should be no issue with setting your sample type and level in the batch table, creating a new method from acquired data, creating levels…
Hello HowngChan ,
For a dual needle multisampler the purge timer does not start until after the method with a different gradient or solvent system is loaded. By default, the worklist will have a 10 minute Wait Time for Ready. This can be set to a maximum…
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x and 5.x
Issue
ICP-MS Instrument Control or Data Analysis Software will not open, or crashes shortly after opening.
Background
MassHunter uses configuration files to store...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.4 and above (Rev C. and Rev D.)
Issue
The Glitch Resetter resets configuration setting files for MassHunter and can be used when the MassHunter software is exhibiting strange behavior...
This Information Applies To: Agilent ICP-MS MassHunter Software 5.x (Rev D)
Issue
It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading...
This Information Applies To: Agilent ICP-MS MassHunter Software 4.x (Rev C)
Issue
It is best practice to note MassHunter Software Settings before running the Glitch Resetter, renaming MassHunter configuration settings files, or before upgrading...
This Information Applies To: Agilent ICP-MS MassHunter Software revisions 4.x and 5.x
Issue
How to clear or archive the acquisition queue? How to resolve queue task file errors?
Background
The Acquisition Queue ( Figure 1 ) lists all tasks...
This Information Applies To: Agilent Fluorescence Detectors G1321A, G1321B, G7121A, G7121B
Issue
This article describes how to diagnose a leak in your fluorescence detector (FLD) module.
Steps to Follow
Confirm the leak error originated in...
This Information Applies To: Agilent ICP-MS MassHunter 5.1 Software, SPS 4 Autosampler
Issue
The SPS-4 autosampler (ALS) may need to be reconfigured after the Glitch Resetter tool is used to reset the software or after a change from a different...
This Information Applies To: Agilent G4208A Instant Pilot
Issue
This article describes how to switch the illumination of the autosampler on or off (if available) using the G4208A Instant Pilot.
Items Required
G4208A Instant Pilot for Agilent...
This Information Applies To: Agilent 7683B Automatic Liquid Sampler
Issue:
O nly use this alignment procedure if the injector hardware configuration has changed or if the Align Mode light comes on.
Agilent does not recommend performing this procedure...
This Information Applies To: Agilent MassHunter data acquisition software and MassHunter Quantitative Data Analysis Software
Issue
This article explains the difference between quantifier and qualifier ions and provides some general guidance on quantifier...