We have a baseline that gives a sinusoidal wave that starts about 16 minutes. We are using an Agilent 6850 GC with a split/splitless injector and an FID. The inlet is interfaced to a 7679A HS. The baseline is good up until 15.5 minutes then it starts…
Hi again
try the last hope ,
first try to select your reporting method in ( DA report template ) , the sequence accept it but i don't know if it will success to print ,
you may try it
the second thing if the first failed ,then try to edit your…
H ztarbs,
The air would not be for an inlet supply. This is most likely a detector support gas for FID. Check your method to make sure that the FID is not on in a standby method. FID air consumption is 400 to 450ml/min when on.
Regards
James
Hello,
We are having an issue with our 7890B GC back inlet. The inlet temperature and flow were both set to "OFF," but the back inlet kept heating up while were were trying to run the front inlet (in GC-FID mode). This resulted in a thermal shutdown…
We've been having issues with our 7890B gas chromatograph. We have it connected to a workstation running Windows 7 SP1 and OpenLab/Chemstation version C.01.07 [27]. We have a Gerstel MPS attached as a headspace sampler.
We have been trying to run…
We use Agilent NGA Analyzer GC based on 4 column configuration. Hayesep Q pre column, Hayesep Analytical column, molsieve column for TCD and one capillary column for FID. The H2S elution takes place in the Hayesep analytical column. But now a days our…
Hello, all you need to do in the method is to uncheck each of the FID flows, and uncheck the electrometer and uncheck the Flame on. With everything unchecked for the FID settings the software will ignore your FID. Under the Signals section you will also…
Hello blairchall ,
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was…
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx analysis software
Issue
When a sample name starts with zero, the leading zero disappears from the exported data ( Figure 1 and Figure 2 ).
Background
Removing the leading...
This Information Applies To: Agilent 7890B GC system and 7697A headspace sampler
Issue
The 7697 Headspace Sampler can be configured to communicate directly with the 7890B GC. When this direct communication is configured, the 7697A headspace gets...
This Information Applies To: Agilent GC/MSD systems
Issue:
When performing MS tuning, there is error message "No Emission Current" and tune is aborted.
Steps to Follow
This error normally means that the selected filament is not working. Perform...
This Information Applies To: Agilent OpenLab CDS
Issue
Any desired processing method can be assigned to the current Result Set.
Requirements
A result set available to link the method to.
Correct Privilege (Role Membership) set up for the...
This Information Applies To: Agilent LC and CE Systems compatible with Lab Advisor
Issue:
Firmware is updated or downgraded when new features are added, or to fix some system bugs. It also solves compatibility issues when a new module is added to...
This Information Applies To: Agilent OpenLab CDS
Issue
This article shows how to export a report of peak area and retention time of a sequence in an Excel format.
Prerequisites
To be able to carry out the procedure as described, privileges...
This Information Applies To: Agilent OpenLab CDS
Issue
Reports in OpenLab CDS can be printed or saved manually in Data Analysys, or automatically during reprocessing, or acquisition.
Requirements
The user needs the following Roles and Privileges...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD systems.
Issue
During tuning, MassHunter prompted the error message: "MSD Fault- RFPA Difficulty", and tune action stopped.
Steps to Follow
Try to reconfigure the GC/MS....
This Information Applies To: Agilent OpenLab CSD 2.5 and above
Issue
The integration optimizer helps find a suitable set of integration events. It allows integration parameters to be tuned via sliders (peak width, slope sensitivity, height reject...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue:
This article explains how to install the Agilent 2100 Expert software and connect a PC to an Agilent 2100 Bioanalyzer system.
Steps to follow:
The latest version of...