Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Hello Samehalsharawy ,
Can you please elaborate on the statement we uninstalled the cylinder expecting the flow to come down to zero but it didn't. It stayed at 50?
If no physical tank is connected to the GC, the GC gas connections are capped…
A compilation of our favorite resources to help you get started with your new GC or GC/MS System:
Free Agilent Education courses are available to guide you through site preparation of GC systems including systems with a Mass Spectrometer, Headspace…
Here's what's new in the latest release of StreamSelect software:
Add support for LC/TQ instruments: Ultivo, 6475A, and 6495D
Add support for PAL3 Series II injector
Add support for license management
Licensing enables sales tracking and to…
This Information Applies To: Agilent 4200, 4150 and 2200 TapeStation.
Issue
Is it possible to analyze single stranded DNA (ssDNA) on the TapeStation? Which kit is the most suitable for analysis of ssDNA?
Resolution
Agilent does not offer any...
This Information Applies To: Agilent 4200 and 4150 TapeStation systems
Issue
How to Run the software installation verification on 4200 and 4150 TapeStation systems.
Background
The 4150 TapeStation model G2992A and 4200 TapeStation model G2991B...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, and Femto Pulse
Issue
Perform a capillary alignment when a new array is installed in your Agilent Parallel Capillary Electrophoresis instrument. An alignment may also be performed...
This Information Applies To: Agilent 4200 and 4150 TapeStation
Issue
Are the Agilent DNA and RNA ScreenTape assays denaturing?
Resolution
The DNA ScreenTape assays in the TapeStation portfolio are not denaturing. Only the RNA and High Sensitivity...
This Information Applies To: Agilent LC pumps with manual purge valve.
Issue
A good practice before using a system is to purge the system to remove air. This is done by observing the clear tubing from the solvent bottle as well as the waste line...
This Information Applies To: ICP-MS MassHunter 4.x and 5.x
Issue
Unable to open a batch in offline data analysis, or the batch is empty.
If the ICP-MS MassHunter software closes unexpectedly, there is a power outage, or the ICP-MS and computer...
This Information Applies To: Agilent OpenLab CDS versions 2.4 or higher.
Issue
How to preview result files using various Report Templates in OpenLab CDS versions 2.4 or higher.
Requirements
The user has the result set loaded in OpenLab CDS...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data ON.
Issue
An 'Authentication failed' error message appeared after entering user name and password when attempting to download a...
This Information Applies To: ICP-MS MassHunter with Spectroscopy Database Administrator (SDA) and Protect Local Data On.
Issue
With Spectroscopy Database Administrator (SDA) and Protect Local Data turned on, batches are not accessible through file...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Scratches on AriaMx optical modules can lead to a reduction in their performance.
Background
Scratches on an AriaMx optical module are shown in Figure 1 .
Figure 1 ....