Hi again
try the last hope ,
first try to select your reporting method in ( DA report template ) , the sequence accept it but i don't know if it will success to print ,
you may try it
the second thing if the first failed ,then try to edit your…
H ztarbs,
The air would not be for an inlet supply. This is most likely a detector support gas for FID. Check your method to make sure that the FID is not on in a standby method. FID air consumption is 400 to 450ml/min when on.
Regards
James
Hello,
We are having an issue with our 7890B GC back inlet. The inlet temperature and flow were both set to "OFF," but the back inlet kept heating up while were were trying to run the front inlet (in GC-FID mode). This resulted in a thermal shutdown…
We've been having issues with our 7890B gas chromatograph. We have it connected to a workstation running Windows 7 SP1 and OpenLab/Chemstation version C.01.07 [27]. We have a Gerstel MPS attached as a headspace sampler.
We have been trying to run…
We use Agilent NGA Analyzer GC based on 4 column configuration. Hayesep Q pre column, Hayesep Analytical column, molsieve column for TCD and one capillary column for FID. The H2S elution takes place in the Hayesep analytical column. But now a days our…
Hello, all you need to do in the method is to uncheck each of the FID flows, and uncheck the electrometer and uncheck the Flame on. With everything unchecked for the FID settings the software will ignore your FID. Under the Signals section you will also…
Hello blairchall ,
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was…
Hello,
I am getting the following error on OpenLab CDS Chemstation Edition (Rev. C01.10 [287]): "User Change (Unlock) failed, because instrument access is not permitted for user. Logging Out User". We are initially able to login once, but then when…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...