Hello,
The test and measurement equipment went to Keysight when it was spun off from Agilent in 2014. You would need to check with Keysight for information on that product.
Marty
Agilent Technologies, Inc. - Agilent Technologies Spins Off Its…
Ok, this indicates a network problem.
Please check the energy settings from windows, that the network card is never going to "sleep".
Also check, that no backup process is running, when masshunter is collecting data.
Hi Stephpal, or sometimes it could be as simple as connecting to ML OQ or ML PC with the instrument powered off; those 2 applications need the instrument to be online and initialized.
Here is more information on Peak Evaluation!
https://www.youtube.com/watch?v=XL0SP-T0CtQ&pp=ygUXYWdpbGVudCBwZWFrIGV2YWx1YXRpb24%3D
https://www.youtube.com/watch?v=GOXghDxtVxI&pp=ygUXYWdpbGVudCBwZWFrIGV2YWx1YXRpb24%3D
If you want to learn more about…
Hello,
I would open a ticket with Agilent informatics support if your installation is client server based or your local Agilent support organization for workstation installation. They can typically help with issues like this where the data needs to…
Hello nikosmpik ,
You will need to go to your WIndows Services app and restart the service, make sure you set that to a Running status, and an Automatic startup type.
To get to the Services app you can do one of two things:
1. Press the Start…
Hello,
You need onsite visit. According to error 60304 and filament time this look like the power tube is too old and end of power life and need replacement.
Hi Metals.
I'm glad to see that you could solve the problem! And I'll take special note since our octopole has never been cleaned. We were told by Agilent that cleaning is meant to be done by Agilent's staff only. Thanks.
Yes, that solution of Ca…
Hi, the peak area has no influence on the calculation of Mw when measuring with concentration detectors only (RI, UV, ELSD e.g.). When performing Viscometry & Lightscattering measurements, the peak area is needed for the calculation. As an additional…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...