Hello AJohnson13 ,
Lab Advisor can be currently downloaded from this page.
Lab Advisor Software Download | Agilent
This may change over time. You can search the web for 'agilent lab advisor download' to find a link to the current download location…
Hello,
So, you should be able to add another column to your report table and change the value to = Compound_Concentration * Sample_Amount / 100 to remove the sample amount and percentage. If you need to remove a multiplier or dilution you will need…
Hello,
LIMS ID is not available as a token for sample name, datafile name, or in CDS DA for report output. You can see the options in the screenshot below. It can be a value in the CSV file but not used in the name. The only other option would be…
Our G4220A (1290 Bin Pump) is setup to work with the seal wash currently. It operates with our instrument with no issue. If I go into the instrument configuration and check the box to enable the pump seal wash (The pump is installed), it will lose connection…
Hi RDP0314,
One thing to check is the injection settings. This behavior is common when the GC is expecting a series of manual injections. To check, click on the Instrument menu shown on your screen capture above and 'Select Injection Source' or similar…
The GPC/SEC Columns Selector Tools help you navigate to the best column choice for your application. Guidance can be by mobile phase, application, manufacturer’s part number cross-reference, USP designation, or EP monograph.
Hi, I always use P/A factor adjustment under Acq/Setup. That works very well. In all your calibration samples a p/a factor adjustment is done. Takes only a few seconds extra.
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...