Hi, we're trying to set up a used 7200 and everything is working but the 7890A we have doesn't have the dedicated CC EPC - is it possible to configure a standard Aux EPC or PCM in a 7890A to control collision cell flow?
Thanks
Durjoy123 would you please start a new forum post with your questions? Please include as much useful information as possible - instrument model number and age of the system, MassHunter Acquisition software version, which high vacuum pump type, the vacuum…
Hello,
I am going to move this post to the GC forum as they will have a better chance of helping. I assume you want to know how to change the front panel language to some other language.
Marty
Hi Kenforce,
In this case, I would run the pressure decay test, as you could have a small leak that is causing the inconsistency. I would also consider increasing your oven equilibration time by a 0.5min or so, just to ensure the column phase is at…
Hi Marty,
We have a FID and an MS connected to the same system. We only want to acquire the signal of FID, but if we want to report the method parameters we are “forced” to acquire both (we do not really know why and we have already ask the question…
Sorry for the delayed response. I've found that either works fine. I typically try to keep injection volume low (0.2-1 uL for 1uM protein). I've mostly been using 50mM HEPES 150mM NaCl and get great deconvolution results, even with a 3 min short method…
I am getting message "draw command failed, EE34205" while running batch in G7129B infinity 2 autosampler. I have tried replacing transport assembly, but error is same. I have used another module G7129A Infinity autosampler with this LC, but same error…
This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared...
This Information Applies To: Agilent Single Needle Autosamplers
Background
Sometimes the user may observe liquid flowing from Port four on the autosampler injection valve. During an injection sequence, it is normal to have liquid from port four...
This Information Applies To: Agilent 4200, 4150, 2200 TapeStation systems
Issue
Can I run an assay using the ladder or sample buffer from a different assay if I have run out of the appropriate assay ladder or sample buffer? Is it possible to run...
This Information Applies To: Agilent ICP-MS 7700, 7800, 7850, 7900, 8800, and 8900 systems
Issue
Peristaltic pump tubing that has lost its elasticity will not deliver the sample to the nebulizer correctly. When the pump tubing appears limp or flattened...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
AriaMx touch screen displays error code 90 after the Instrument Health Check (IHC), which happens when the AriaMx is switched on, or after running the onboard diagnostic tests...
This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx optical modules
Issue
AriaMx users can change which AriaMx optical modules are installed to upgrade the instrument capabilities or to troubleshoot.
Items Required
Any...
This Information Applies To: AriaMx real-time PCR system
Issue
Error 43 is displayed on the AriaMx touch screen. Error 43 is a voltage error and it normally indicates that the cartridge LED voltage or motor voltage being detected is abnormal. Sometimes...
This Information Applies To: Agilent 2100 Bioanalyzer system
Issue
Can I run qPCR samples containing Sybr® Green I on the Bioanalyzer?
Background
Sybr® Green I is an Intercalating Dye commonly used to detect amplification products in real...