Hi, we're trying to set up a used 7200 and everything is working but the 7890A we have doesn't have the dedicated CC EPC - is it possible to configure a standard Aux EPC or PCM in a 7890A to control collision cell flow?
Thanks
Hi, we're trying to set up a used 7200 and everything is working but the 7890A we have doesn't have the dedicated CC EPC - is it possible to configure a standard Aux EPC or PCM in a 7890A to control collision cell flow?
Thanks
No. There is a QTOF specific collision cell EPC module. The software needs to recognize it.
G3850-60039 | Agilent EPC module
G3470-60550 | Agilent, Fitting and 1/16" tubing to go from the EPC module to the QTOF CC 1/16" Swagelok connection.
Is the 7890 setup for the MS transferline heater/sensor connection? Has that GC ever been run with a mass spec?
Is it a 7200A or 7200B ? What's the serial number?