Offsetting the axes, Y or X, is not a feature in MassHunter. You can do head to tail comparison.
Overlaying in different colors and choosing which one is active can be done but there's not a way to do what you're asking even in a custom report.…
Good morning. We analyze fatty acids and The compound response factor of each fatty acid before the softwere changed remained fixed now it changes. I tried to copy the various parameters set in the other gas chromatograph which has its own softwere, but…
Hello,
I'm having the same problem with the overwhelming amount of orange highlighting for samples where the qualifier peak was not found. I tried the suggestion above of enabling "Ignore Peaks Not Found" in the method and re-applying it to the batch…
The tune evaluation was written for an orthogonal ion source (SS/Inert/Extractor) system, atune or etune, running helium carrier gas between 0.8 and 1.4 ml/min, at 230° C ion source and 150° C quad temperatures. It may not pass using other temperatures…
Hello,
I am not sure which Agilent software (CDS 2.x , Chemstation, EZChrom) you are using, but most customers use third party software for simulated distillation and other RT vs BP petrochemical calculations. If your need is simpler, you could do…
Thanks a lot.
Also I found that not all parameters (column width, labels) can't be changed in report designer. I changed them in xml template manually. Is there any description of functions and variables to change some report layout?
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...