The Deuterium ( D2 ) UV lamp disposal depends on the local regulations, which are different in every country. The D2 lamps (bulb is under vacuum) only contain a fairly small amount of Deuterium, which is not harmful. May be considered industrial waste…
Hello afzishle ,
I apologize for the delayed response.
It is possible to limit the number of results displayed in Report Builder tables or lists. You would sort by whatever you are evaluating your results by and then at the end of your filter use…
I'll just guess because you gave us very little information in my opinion.
But...It seems m/z 208 is shifted to more cc than it should. Maybe an interference with that m/z?
The 66x2mL tray has three white square plastic clips on its rear for automated identification by the sampler.
The 18x6mL tray just has one of the squares.
Please check your tray, i am pretty sure that some of these clips are missing.
You may will…
Okay, I just remembered - I didn't manage to change the size of the text in this program, it was in the Qualitative analysis program. So perhaps it is not possible to change the text size in this program. I really wish that feature would be added, for…
Hello Chalbrecht ,
I apologize for the delayed response.
There are a few options for naming the reporting output file using the sample or batch name with $(SampleName) and $(BatchFileName), but I don't see that the Sample Group is an option. You may…
Hello Achim_Hongenaert ,
I apologize for the delayed response.
There are a few outliers that can be used for recovery calculations. If this compound is added to all samples then I believe the best option will be to call it a Surrogate in quant, even…
This Information Applies To: Agilent Atomic Absorption Spectroscopy Systems (240/280 Series Flame AA and 240/280Z/Zeeman/GTA 120 Graphite Furnace)
Issue:
This article includes recommended maintenance procedures to be performed daily, weekly or annually...
This Information Applies To: Agilent SpectrAA Software, Agilent 240/280 Flame Atomic Absorption Systems, 240/280 Zeeman Graphite Furnace Systems
Issue:
How to Reset Lamp Usage Counters or Add a New Lamp in SpectrAA
Caution : All physical...
This Information Applies To: Agilent GTA 120, Agilent 240Z AA Spectrometer, Agilent 280Z AA Spectrometer, Agilent 240FS/GTA120/PSD120/UltrAA System
Issue:
It is important to align the sample probe of the Programmable Sampler Dispenser (PSD) 120...
This Information Applies To: Agilent Flame AA Systems: e.g., 55B, 240FS, 280FS
Issue:
The Flame AA burner and burner components require regular cleaning to maintain system performance. Components include the burner, nebulizer, and spray chamber...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to clean a blocked nebulizer of a Flame AA, which may be causing low sensitivity or a sample uptake rate below 5 ml/min.
Steps to follow...
This Information Applies To: Agilent Graphite Furnace Systems (PSD 120/GTA 120/240Z AA/280Z AA)
Issue
Maintenance procedure for replacing the syringe and plunger of the PSD 120 sample dispenser for the Agilent GTA and Zeeman Graphite Furnace Systems...
This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)
Issue:
The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.
The PSD enables the...
This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z
Issue:
Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the...
This Information Applies To: Agilent VGA 77, AA, VGA 77P, ICP
Issue:
The Agilent Vapor Generation Accessories VGA 77 and VGA 77P use high concentrations acids. The system needs to be thoroughly cleaned after each use.
The Vapor Generation Accessory...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ...