Hi, I print all my reports with choose samples, select by group. Is there a way to include the sample group in the file name?
This is for MassHunter10.
Hi, I print all my reports with choose samples, select by group. Is there a way to include the sample group in the file name?
This is for MassHunter10.
Hello Chalbrecht ,
I apologize for the delayed response.
There are a few options for naming the reporting output file using the sample or batch name with $(SampleName) and $(BatchFileName), but I don't see that the Sample Group is an option. You may be able to use a Post Process command that would prompt you for the group name and then rename the output, but without extensive work I'm not certain there is a way to do this directly from report generation.
Hello Chalbrecht ,
I apologize for the delayed response.
There are a few options for naming the reporting output file using the sample or batch name with $(SampleName) and $(BatchFileName), but I don't see that the Sample Group is an option. You may be able to use a Post Process command that would prompt you for the group name and then rename the output, but without extensive work I'm not certain there is a way to do this directly from report generation.