Hello,
Are you sure the end tip of the PLOT column ends up (and 2 mm withdrawn) inside the FID Jet?
Any problems when igniting the FID?
FID Jet could be contaminated from PLOT stationary phase particles ..
Best regards -
Lars
Our lab has an Agilent 1260 Infinity HPLC, and we are encountering an annoying issue with the desktop connected to it. The desktop sometimes shut down automatically, while the HPLC is still running the analysis. The HPLC keeps working and and the analysis…
Hello, anyone here know what happen with my icp-oes?
i got error message '' gas scp write channel" and icp-oes unable to running due plasma ignition failed, the icon plasma also blinked many times.
Thank you.
Rerey
The vials and caps are the same ones we have been using for years now; they are your standard 2mL GC/LC vials. I opened up LabAdvisor and ran the gripper test. It failed to release the vials even after re-initialization. I removed the gripper caps, wiped…
Here is a collection of Agilent 7890, 8860 and 8890 GC Resources. Links to user manuals, troubleshooting resources , tools and supplies on Agilent.com . Go to Collection of GC/MS Resources to view GC/MS information.
Agilent 8890 GC User Manuals
…
Hello,
I am new to MassHunter, so this may be a basic question or just something I'm doing wrong. I am trying to build a quantitation method for a large number of compounds we are analyzing via LCMS. I ran our calibrators with our LCMS method and have…
Hi
When FID is off means no flame.
Voltage applied during ignition and while FID OFF so ignitor also Off..
When ignitor off then voltage also off...
Hope, it's clear to you..
The Pfeiffer turbopump on my 5977C MSD will not start after shutdown; it does not exceed 5%. When I boot the MSD, the green LED on the pump controller is solid. Then when the turbo fails to reach 80%, the LED starts blinking and the MSD panel returns…
If your area is relatively stable but your amount/result is low, you have to have either altered your calibration values, failed to clear a previous calibration, or omitted a dilution factor. Are all results lower by the same factor or is it variable…
This Information Applies To: Agilent 4200 TapeStation 4150 TapeStation, TapeStation controller software
Issue
The system verification can be run to check the TapeStation performance, for example, when you suspect an instrument hardware fault.
...
This Information Applies To: Agilent 4200 and 4150 TapeStation software.
Issue
How to uninstall and reinstall the 4200 and 4150 TapeStation software.
Background
An uninstallation and reinstallation of the TapeStation software should be performed...
This Information Applies To: 2200 TapeStation instruments, 2200 TapeStation controller software
Issue:
The Agilent 2200 TapeStation controller software won't launch but displays the correct serial number in the top-right corner of the loading screen...
This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher.
Issue:
When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, ProSize Data Analysis Software
Issue
How to export a ladder as a size calibration file and import it into a different run in ProSize.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
It is possible to set up the 2100 Expert software so that the files are automatically exported each time a chip run is complete.
Steps to Follow
To turn...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
How to identify and troubleshoot late migration of the ladder, marker, and sample peaks in data generated on the Agilent 2100 Bioanalyzer system.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
Instructions on how to import from or export data into LIMS software.
Steps to Follow
Data import
When importing sample names and sample comments from...