This look like it's only bad tuning parameters. Try to have another tune file for EPA. Also, you are doing dilution with EPA. You should start with the wizard method to see if you can fix this. it has the tuning parameters. You don't need to perform resolution…
You can expand the details of the error message. This could indicate a problem with the HDMI cable. It might need to be replaced or reseat. If you have the details of the error that could help.
Hello,
From the error code this look like error 5 with red light 2 and 4 on. Axis movement failed. Either something is blocking the X movement or there is a problem with the sensor. This will require repair. Please call the response center to get…
The main error is 1220. you have to focus on this. The other error are consequence of 1220. Make sure you don't use higher than 10% IPA or it will clogged the cones. You need Argon/Oxide mix to remove the carbon from the cone tips if you use organic solvent…
Hello,
Help about to know what version you have. If it's masshunter you have the option to export with transposition. Or menu file, export. I can't recalled the ChemStation revision has everyone was upgrade to latest version of 7500 software. B.01…
Lithium-ion batteries (LIBs) have emerged as indispensable power sources in a range of applications. With the steep rise of LIB production (and recycling) comes the need for rigorous quality testing of the components used in these batteries.
The Agilent…
Thank you very much.
We actually monitor mass 56 (among with others) at the beginning of each session because mass 56 counts start at around 30-70 K (skipping warm up, just when the plasma lights up) and go down till ~500 counts in around 20-40 minutes…
For WinUV 5.1.3.1042 SCM has an E-sign option, which says to go to the profile screen which does not have 'Authorize Worksheet' and 'Review Worksheet' Privileges to select….Kind of a wild goosechase.
Click 'Configuration > Electronic Signature' to access…
Here are the services I suspect are not running after your last reboot. Another reason is that the graphing options are available and graphics and annotation are not from the SDA.
This Information Applies To: Agilent 4200 TapeStation 4150 TapeStation, TapeStation controller software
Issue
The system verification can be run to check the TapeStation performance, for example, when you suspect an instrument hardware fault.
...
This Information Applies To: Agilent 4200 and 4150 TapeStation software.
Issue
How to uninstall and reinstall the 4200 and 4150 TapeStation software.
Background
An uninstallation and reinstallation of the TapeStation software should be performed...
This Information Applies To: 2200 TapeStation instruments, 2200 TapeStation controller software
Issue:
The Agilent 2200 TapeStation controller software won't launch but displays the correct serial number in the top-right corner of the loading screen...
This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher.
Issue:
When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, ProSize Data Analysis Software
Issue
How to export a ladder as a size calibration file and import it into a different run in ProSize.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
It is possible to set up the 2100 Expert software so that the files are automatically exported each time a chip run is complete.
Steps to Follow
To turn...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
How to identify and troubleshoot late migration of the ladder, marker, and sample peaks in data generated on the Agilent 2100 Bioanalyzer system.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
Instructions on how to import from or export data into LIMS software.
Steps to Follow
Data import
When importing sample names and sample comments from...