I apologize for the delayed response.
Errors of this type generally point to an issue with a preferences file. The first thing to try is clearing these using the Restore User Settings tool found in the Quant Tools group in the start…
Hello AJohnson13 ,
This is a known issue if Acquisition has been pinned to the Windows Taskbar and run from there in MassHunter Acq Version 11.x and later. This is due to the OpenLab integration and the need to launch through Control Panel to include…
Please give more information. TSU - is that Thermal Screening Unit? You say under varying pressures...from zero to how many PSI/Bar? Will you need to use a vacuum to get the sample? Will this be analytes in air or something else? Decomposition of what…
Did the EM get replaced March 15? Or a source cleaning? The EMV went down 883V.
The Ent Lens (Entrance Lens) those jumps from 16 to 63,32,47.5 .. and the Ion Focus going from 80's to 40's... make me jumpy. Something was changing. When were source cleanings…
The ID testing process in the pharmaceutical industry often exposes warehouse staff to hazardous materials. This new technical note shows how pharmaceutical manufacturing QA/QC organizations can use the Agilent Vaya Raman system to check raw materials…
Hello community,
I have the following error (attached images) when start either offline or online mode. As you can see, the module control panel also dissappear. I have checked all the conections between PC and all modules and the error window continue…
Recommendations updated after looking at this version from February 16, 2023. The recommendations are pretty much the same as above.
Increase the oven temperature equilibration time to 1.0 to 1.5 minutes or longer. It takes time for the 240° oven post…
We attempted the same testing on another OL2 instrument and the sensitivity issue was not present. It was unique to a single system. Agilent was very helpful getting us another system to test on. It seemed to be related to the tune/threshold/filtering…
Hi,
1. This is the "Easy Fit Lense Assy" - Since two years, we (Agilent) has decided not to clean the lenses, but to replace them completely. So if you want to disassemble these, you have to break the plastic cap, and use a screw set from an older regular…
This Information Applies To: Agilent 4200 TapeStation 4150 TapeStation, TapeStation controller software
Issue
The system verification can be run to check the TapeStation performance, for example, when you suspect an instrument hardware fault.
...
This Information Applies To: Agilent 4200 and 4150 TapeStation software.
Issue
How to uninstall and reinstall the 4200 and 4150 TapeStation software.
Background
An uninstallation and reinstallation of the TapeStation software should be performed...
This Information Applies To: 2200 TapeStation instruments, 2200 TapeStation controller software
Issue:
The Agilent 2200 TapeStation controller software won't launch but displays the correct serial number in the top-right corner of the loading screen...
This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher.
Issue:
When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, ProSize Data Analysis Software
Issue
How to export a ladder as a size calibration file and import it into a different run in ProSize.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
It is possible to set up the 2100 Expert software so that the files are automatically exported each time a chip run is complete.
Steps to Follow
To turn...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
How to identify and troubleshoot late migration of the ladder, marker, and sample peaks in data generated on the Agilent 2100 Bioanalyzer system.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
Instructions on how to import from or export data into LIMS software.
Steps to Follow
Data import
When importing sample names and sample comments from...