I made a custom excel report template for GC-MS/FID data I'm working with in MassHunter B.10 and I'm having an issue getting the template to report the user selected "best" from the spectrum ID results list.
Here is how I've added the table…
Before first use: Soak in 100ppm sodium calibration standard for 24 hours to condition the probe With use (submersion in the liquids under test) the response time will become slow and it might be difficult to calibrate. If this occurs the probe needs…
Thanks for the response Matthais. The issue actually is that autotune sometimes fails to match EPA compliance criteria, and so it becomes necessary to manually tweak settings. I get that Agilent doesn't want lab techs carelessly messing around with quadrupole…
Hi! Recently we baught GC 8890 with tandem detector fid-scd(detector for sulfor detection 8355). In this case limit of detection of sulfur compaunds is very high (ppm level). We need to meet more low limits of detection of sulfur compounds. Can we dismount…
Thank you sir for the best answers.
I have another question to ask.
My new MPAES 4210 is just a year old. My plan is to Analyse metals for Lube Oil Condition Monitoring for mine fleets.
However, I find bit difficult to get the correct calibration…
Abdoul-
From your description (instrument configuration fails), it sounds like you will need to remove and re-install MH Acquisition. You might try a repair of the software.
Pinging 192.168.254.12 only confirms that the MS is communicating with…
I am using MassHunter Qualitative Analysis B.07.00 to analyze FID data. I have extracted the FID chromatograms, identified the peaks and annotated them. But nothing happens when I try to print reports. Can anyone help me figure out how to generate the…
Thank you very much! I'll make sure to have those items correctly. Our instrument (5975C MSD) has been in idle condition for over an year but under vacuum up until recently I started it again. My real issue is autotune fail. The tune fail symptom is Ent…
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...