This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
Items Required
Steps to Follow
Ensure that the ISTD and sample peristaltic pump tubing tension are set properly
Figure 1. Peri Pump Tubing Tension Diagram
1. Proper tension, 2. Tight sample line and loose ISTD line, 3. Loose sample line and tight ISTD line
Maximize internal standard and sample mixing
![]() |
![]() |
Advantages:
|
Advantages:
|
Disadvantages:
|
Disadvantages:
|
MicroMist U-Series Nebulizer Sample Tubing 0.25 mm id (p/n G3266-80013) |
MicroMist U-Series Nebulizer Sample Tubing 0.5 mm id (p/n G3266-80012) |
Table 1. Internal Standard Mixing Setup Comparison
Prevent dead volume issues in the connection block
When the autosampler probe raises out of the sample tube, air is taken into the connection block via the sample line. This can lead to random high RSDs for the internal standard.
Tip: This is applicable for non ISIS3 configurations only!
|
Figure 2. ICP-MS Instrument Control Dashboard
1. Sample Introduction, 2. Special Process for Sample Change
Figure 3. Special Process for Sample Change
1. Stop PeriPump Rotation, 2. OK button
Figure 4. Batch Sample Introduction (PeriPump/ISIS) Page.
1. Preemptive Rinse, 2. Save Batch button
Learn how to effectively troubleshoot your Agilent ICP-MS System: Agilent 7800/7850/7900 ICP-MS with MassHunter Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning courses available from Agilent education. |