This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated hours on the UV lamp.
Chromatography Data System software
Figure 1. Lamp Tag Information
|
Lab Advisor
Figure 2. Hours on UV Lamp via Lab Advisor
1. HPLC connected to Lab Advisor, 2. EMFs, 3. Accumulated UV Lamp On-Time
Instant Pilot
Figure 3. Welcome screen on Instant Pilot
1. More, 2. Maintenance
Figure 5. Maintenance screen of DAD on Instant Pilot
1. Maintenance, 2. Lamp Info
Infinity LC Companion (Available in selected regions)
Figure 7. Control screen on Infinity LC Companion
1. Control, 2. Lamp Info
Figure 8. Lamp Info on LC Infinity Companion
|
Confirmation: The hours will be set to zero when a new lamp is installed. No reset maneuver is needed.
|