Hi Lance.
Setting the FID to 5 Hz fixed the peak noise issue. I did notice however that when lowering the makeup flow on the FID < 5 ml/min (set it at 3 ml/min) there was loss of sensitivity in the FID. I set it back to 25 ml/min and the sensitivity…
Dear Colleagues,
I have GC-FID 8890 and i'm using Gas saver feature but I'm observing when I change time of gas saver , the retention time is changed.
for instance, I set gas saver at 20 ml/min after 2 min then inject methanol and the RT is 4.74 but…
The boxes of standards should have shipped with a Certificate of Analysis enclosed. If you go to Agilent.com, you can search on the part number. A number of the hits are CofAs from different batches. part 5080-8858 CofA Lot CF-4631W.pdf (agilent.com)…
Ah, and I think the issue has to do with that I have 2 autosamplers, and 3 columns (MS at back, and 2 different FID columns at front using a 2way splitter). So the MS signal is in the subsequent folder, the FID1 signal is in the correct folder, and FID2…
Hello,
To calculate the RSD and Pass/Fail results, one can try the following:
In the reporting go to the edtor and open "table properties"
In my specific case, I want to calculate RSD for Area, so I choose the column A, pA*s and press "Column…
Hi everyone, My MP-AES 4210 Machine sa is failing to suck/take-up the samples from the sample tubes..
I tried changing the sample pipes but nothing worked. Kindly asking for help on how to sort out this problem?
Regards,
Kalunga
Hello!
It would be very good to get step by step instruction and definition of each row on Splitter calculator and other calculators to configure our system.
The system uses SCD and FID detectors.
Target is to get ratio 3:1 between SCD and FID for…
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...