Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...
This Information Applies To: Parallel capillary electrophoresis platforms, 5200/5300/5400 Fragment Analyzer, ZAG DNA Analyzer, Femto Pulse, Oligo Pro II
Issue:
Disposing used capillary arrays
Resolution:
Arrays can be discarded in regular...
This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, ProSize software, DNF-471 RNA kit, DNF-472 HS RNA kit
Issue:
The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE...