This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, ProSize software, DNF-471 RNA kit, DNF-472 HS RNA kit
Issue:
The DV200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents the percentage of RNA fragments larger than 200 nucleotides relative to all RNA fragments. The DV200 of samples run with Agilent Fragment Analyzer DNF-471 RNA kit or DNF-472 HS RNA kit can be calculated within the ProSize data analysis software.
Steps to follow
This article contains different instructions for 33 cm arrays and 22 or 55 cm arrays.
Running a Separation with a DV200 Method (33 cm array)
The DV200 is provided under the Smear Analysis tab in the % of total column for each sample in ProSize.
Applying the DV200 Metric to a Previously Run Data File (33 cm array)
The DV200 is provided under the Smear Analysis tab in the % of total column for each sample.
Figure 1. Applying the DV200 configuration to an existing data file
Applying the DV200 Metric to a Previously Run Data File (22 or 55 cm array)
Method and configuration files are only available for the 33 cm array. However, If you have a 22 or 55 cm array it is still possible to calculate the DV200 by setting up a smear range within ProSize:
The DV200 is provided under the Smear Analysis tab in the % of total column for each sample.
Figure 2. Smear Analysis tab in the Set Individual Parameters dialog box
|
Learn how to effectively maintain and operate your Fragment Analyzer:
|
For Research Use Only. Not for use in diagnostic procedures.