Hello,
I am expeirencing the same issue after having to reload E.02.02.1431. The quant report prints as one job and the chromatogram and the spectra reports (I'm printing a detailed report) print as a second job.
I've done as ron_tackett mentioned…
Instability in the anode voltage will cause baseline problems in your detector signal. For example, if the lamp doesn't light properly, the voltage will reset and you will see periodic changes in the baseline. I have seen old lamps sometimes produce odd…
I can't seem to get our new instrument to pass an EPA DFTPP tune for method 525.2. We recently got an instrument with the hydroinert source and I have only gotten a couple of passing tunes but nothing that is repeatable. I have tried manual tuning myself…
Yes! I checked the line too as according to the instructions (+) How to Disassemble and Assemble the HES (High Efficiency Source) - Articles - GC/MS Portal - Agilent Community
I will take a picture when I have them out next time. Thank you so much for…
Analyze Agilent Seahorse XF Real-Time ATP rate assay data using Agilent Seahorse Analytics
Analyze Agilent Seahorse XF Hu T Cell Activation Assay data using Agilent Seahorse Analytics software
Analyze Agilent Seahorse XF Substrate Oxidation Stress…
Hello APacana ,
The reason for the difference in response between the TIC and the EIC is because when you extract the TIC of a given precursor you are extracting the ion signal from all precursors that create a 72 product ion, using your example. If…
Hello AJohnson13 ,
Here is the PC setup guide for older Windows 10 versions of MassHunter. It includes the PC specifications.
Windows10_Setup.book (agilent.com)
The software installation instructions would be included as part of your original media…
At 1 mL/min pumping water, the pressures are as follows:
Pump + Column = 53 bar
Pump + Column + 0.1 mm ID 30 cm tubing = 70 bar
Pump + Column + Tubing + Flow Cell = 75 bar
I have also confirmed that this pressure fluctuation has nothing to do…
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...