For more information, check out the application note Five Keys to Unlock Maximum Performance in the Analysis of Over 200 Pesticides in Challenging Food Matrices .
Recommended parts from the video:
Self-tightening, collared column nut for use with…
Hello,
Are you still having low flow issues on your AA? Did replacing the nebulizer assembly help?
Do you change the flame gases for your methods? If so, call up a default air/acetylene method. You will need a lamp of that element. Go to Optimize…
It is hard to see from the picture which stator this is. If it is the 7 port for ISIS3 then the part number of just the stator is:
G8411-80121 Stator, for 7 port, 2 position inject valve.
Hello,
The penning gauge is now referred to as the ion gauge. It is located on the right side, front of the analyzer chamber. It is easily identified by the blue cable connected to the back of the gauge. Make sure to put the instrument into Shutdown…
With the rapid move towards electrification, lithium-ion battery (LIB) production is expanding to keep up with global demands. Strict quality control of the materials used in these batteries is paramount for battery safety and performance.
The Agilent…
This single-channel GC offers the built-in intelligence and high performance of the 8890 GC in about half the footprint. The small column oven enables fast thermal cycling, which contributes to a reduction in power usage of up to 30%. Ease of maintenance…
Hello glee168 ,
I never saw that error nor heard of people getting that. Where is that error appearing, on the PC, on the GC display, or is it the tool that is giving that error? Or, were you running the chromatography software while the FW update…
Hi Marty,
I am Yew Mun colleague.
Thanks for your help. we run samples in sequence. The ChemStation version we have is 2.141.0.0 . I will try to setup the report in method and let you know how it goes.
Hema
Hi
I have one problem with my 5977B.
The tune no finish, the principal problem or error is "Tune action stopped: Gain and offset = 0 and peaks are too narrow".
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...