Hello Adamcfc101 Please advise if the intermittent error "metering device command failed, draw command aborted" is still occurring on your dual needle multisampler. As Carlos has asked, please also share the following details:
- is this error happening…
Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Realize this was a while ago but I just came across it; wrote something similar a while back that processes the data in a given sequence folder - could easily be adapted to run as post-run and post-sequence or also with a specified folder. Hopefully this…
Thank you Paul! I haven't got a chance to try it yet. But I found the sample might carry too much salt (in heptane) that may damage the column and system. The drop always occurs right after one of the std peak which lead me to think it is somehow LRI…
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...