This presentation shows basic information, maintenance procedures, and exhaust venting for the three possible rough pumps used on Agilent single quad GC/MS systems.
Rough Pumps on Agilent 5977x GCMS systems May 2021 rev2.pdf
I've been doing 525.2 for close to 16 years, and lately been having issues with perylene d 12, and triphenyl phosphate failing the +/- 30% of targeted concentration. I don't think the issue is due to my sample concentration step, as the nitro xylene recovers…
There is no written maintenance procedure for the MS40+ maintenance kit.
Here is an exploded view of the pump:
Here is the parts breakdown for the maintenance kit:
I've been doing 525.2 for close to 16 years, and lately been having issues with perylene d 12, and triphenyl phosphate failing the +/- 30% of targeted concentration. I don't think the issue is due to my sample concentration step, as the nitro xylene recovers…
Hi David, this option exists already in MH 4.3 and is called Intelligent Sequencing. It is an additional software package, not included in basic MH. You can buy a license key and unlock its features. This is a very powerful software tool which increases…
We're attempting to install MassHunter Workstation Quantitative Analysis Reporting B.07.00 but it keeps failing with an Error: Unable to get the Add property of the AddIns class about halfway through the process. All other software and reporting has installed…
Hi Lorelei, In order for me to fetch correct information could you list model numbers of tower and controller, i.e. 7683 ALS tower. Meanwhile I was checking for different error messages and couldn't find Radius home failed. If error is still present could…
I am currently trying to setup a GC system and am having issues with MassHunter Acquisition. I am trying to setup a parallel system to one that we are currently running. The new system is 6890N/7697A HS, it does not have a MS. I am trying to use MassHunter…
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Over time, the electrodes may become pitted or eroded and users may see a decrease in precision or begin to experience the "Tube fault" error when new tubes are installed. If...
This Information Applies To: Agilent GTA Furnace and Flame Instruments, Agilent 240FS/240GFAA/GTA120/PSD120
Issue
There may be occasions when users need to remove GTA furnace center block workhead (instruments with both Flame and Furnace capabilities...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
The shroud should be replaced annually, or when there is evidence of damage caused by mechanical shock or contamination from liquid spillage.
Items Required
Shroud Removal...
This Information Applies To: Agilent GTA 120, 240/280 GTA Furnace
Issue
Before replacing the electrodes and/or shroud, the GTA Furnace workhead will need to be disassembled. After replacing the electrodes and shroud, the GTA Furnace workhead will...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Check...
This Information Applies To: Agilent OpenLab CDS 2.x in file system configuration
Issue
Sometimes, you will need to change the Projects Root Path / Project Storage Path . This article outlines the procedure on how to change it.
Background
...
This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7
Issue
Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES
Tip : These steps should be completed after the Subsystem Communication...
This Information Applies To: Agilent 8890/8860 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent 6000 Series LC/MS systems equipped with MS40+ rotary vane pumps
Issue
Instructions on how to replace the filter in an Agilent MS 40+ pump as part of annual pump maintenance
Background
The filter in the...
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) with ICP Expert 7.5.x Software
Issue
The Recommended Routine Maintenance Schedule does not adjust for the number of samples analyses or the sample matrix, both of which...