Hello Yanding, Glad to hear that worked! The needle to seat connection is very resilient to pressure, and can hold at 600 bar and more. I do not believe that would be the reason why the connection failed, it is most likely that the screw or other part…
Hi, I would be grateful for the community's insight on an issue that has been plaguing our 6520 Accurate-Mass QTOF-LC/MS. We are able to successful turn on the instrument, power condition, do mass calibrations, and TOF tunes (all the way from quick tunes…
Use 1 meter of deactivated fused silica tubing going into the source (the same diameter of your column), then use a union (either the press-fit glass type of, better yet, the SilTite brand) to connect to your column in the oven. When you want to change…
Hello tdnola , "in use by another..." is a safety function that prevents a second session to the MSD: OK & wanted function. 1st) troubleshooting because a PING does only a small (OSI) communication check (are voltages & cabling OK?), I recommend the communication…
#I am currently battling with MH 4.5 and a generic communication error that has cascaded into me not being able to bring the 7700x into “shutdown” mode for ~6 weeks now. I have had numerous calls with tech support and all suggestions have failed to this…
We work with a Q-TOF 6545 and it fails positive auto tune. Is there anyone who knows this error, what causes it and what can be done to rectify this? Thank you
Hello everyone,
I have created a simultaneous SIM/scan method and we're using a HS-SPME-GC-MS system (GC: Agilent 7890 A GC-System; MS: Agilent 5975C Inert XL MSD with Triple-Axis Detector; Software: Chemstation G1701EA E.02.00.493).
We get a TIC…
Hello, I am trying to generate a streamlined intelligent report in OpenLab where we can capture all peaks of our SEC gel filtration standard in a single table and be able to evaluate APD Pass/Fail for each peak and Resolution between BGG and Ovalbumin…
Were you able to join us on April 11th for our Agilent GC Live session? If so, we would like to hear your thoughts and questions. If you missed it, please see the video below.
Time stamps refer to the above video.
(14:05) What is…
Hi I'm using chemstation revision G1701EA E.02.02.1431 and the last successful launch was a day yesterday. It seems the malfunction accrued (during a "run" with a "new" method we are trying. involves FID HS for DCM residue, all the vials were sampled…
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue:
This article explains how to install the Agilent 2100 Expert software and connect a PC to an Agilent 2100 Bioanalyzer system.
Steps to follow:
The latest version of...
This Information Applies To: 2100 Bioanalyzer systems, 2100 Expert software
Issue: W hen trying to launch the Agilent 2100 Expert software, the process aborts and the user sees run-time error message :
Run-time error '50003 ' often indicates...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software without Security Pack addon
Issue
The 2100 Expert software gives a counter mismatch error in the log book which may lead to data loss (blank lanes). Specifically...
This Information Applies To: Agilent ICP Expert software.
Issue
The software version is often asked and is a valuable information: to see if any updates are needed, if certain options are available etc.
Steps to Follow
Launch ICP Expert...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to calculate peak-to-valley ratio in OpenLab CDS.
Background
The peak-to-valley ratio is calculated to indicate the quality of peak separation...
This Information Applies To: Agilent 6500 Series Q-TOF LC/MS systems, MassHunter Workstation 10.1 or below and MassHunter Workstation 11.0 .
Issue
Fully reset the Q-TOF LC/MS system via a power cycle without venting. For users who experience slow...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II systems
Issue
The volume of sample used by Agilent parallel capillary electrophoresis systems during a run.
Resolution
The...
This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message " Start of launcher application failed - The requested...
This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Preventive or periodic maintenance...