This Information Applies To: 2100 Bioanalyzer systems, 2100 Expert software
Issue: When trying to launch the Agilent 2100 Expert software, the process aborts and the user sees run-time error message:
Run-time error '50003' often indicates that an older version of Microsoft Common Controls might be installed on the PC in question. Possible root causes include:
Computer display (scale and layout) not set to 100%
Remnant of an old Visual Basic program
Files not registered properly during 2100 Expert software installation
Necessary .OCX files cannot be overwritten as another program uses an older instance
Steps to follow:
Check that the scaling of computer display is set to 100 %. Select computer Start menu > Settings > Display > Scale and layout > select 100% (Recommended) from the drop down menu.
Open the 2100 Expert software from the desktop shortcut. If the shortcut does not launch the software, open the software from the installation directory. Select C:>Program Files (x86)>Agilent>2100 Bioanalyzer>2100 expert >bin, hold the SHIFT key and right-click the ExpertLauncher.exe. There will be two options that can be tested:
Run as administrator.
If issue persists, reinstall the software from local administration account with full administration rights. Alternatively, if you are not running the most recent version of the software then update to the latest version of the software under local admin account. Before the software upgrade, please see software Readme file to ensure that PC specifications are met for the most recent version of the software. Then review the software installation steps from the read me file. Briefly, for software reinstall or upgrade, see following steps:
Back up data
Caution: Failure to back up data will cause irretrievable loss of data.
|
Uninstall the software.
Reboot the computer.
Reinstall the software anew
If there are still issues, create a new local administrator user profile and repeat step 3.
Contact your Agilent local support team for further assistance as needed.