This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software without Security Pack addon
Issue
The 2100 Expert software gives a counter mismatch error in the log book which may lead to data loss (blank lanes). Specifically, it occurs when the 2100 Expert software on the PC detects that there are data points missing in the information it is receiving from the Bioanalyzer during a run. This indicates an interruption in the communication between the instrument and the PC.
Background
The Counter Mismatch error is a communication related error. If the data/run otherwise looks fine, it indicates that the number of dropped data points is low, so the results from the run should still be good in spite of the presence of the Counter Mismatch error. The electropherogram is composed of hundreds of data points, therefore, in general, the loss of just a few data points will not cause any noticeable effect on the results. If too many data points are lost, this will result in data loss (blank lanes).
Resolution / Steps to Follow
Review the communication troubleshooting chart on page 19 of the Maintenance & Troubleshooting Guide. Communication interruptions like this could be due to a variety of things.To begin with, address the following:
If the communication issue persists after reviewing and confirming that the above recommendations are being followed, consider the following next steps:
Tips: Learn how to effectively troubleshoot your Bioanalyzer:
Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide |