Yes, the MassHunter software uses a conversion factor and reports in "counts" instead of hertz for uECD data. The conversion factor is 7680.
The 7680 multiplier also applies to other GC based detectors such as FIDs and TCDs.
The basic tools I would recommend to have for GC maintenance:
Ceramic column cutter
1/4" wrench (column nuts)
7/16" wrench (gas fittings at back of GC)
1/2" wrench (replace gold seal on S/SL inlet)
Torx 20 driver (removing FID collector assembly…
What you describe with the washes happening, but a delay when pulling up the actual sample for injection sounds like something about the GC is not in a ready state and the GC is waiting for a part of the GC to become ready. In this case you describe it…
Hello igorvlad
I’m sorry for the delayed reply to your message. Do you still need assistance with this? If you do, please take into consideration the notes below. However, if the issue was resolved, could you please share what the resolution was?…
Hello !
I interesting gc8890 and gc7890.
In datasheet the gc 7890 writing about posible instal 3 detector maximum, but in datasheet gc 8890 writing about posible 4 detector. I never see 4 detectors are installed simultaneously on one chromatograph…
Hi, I analyze methanol in water just about everyday. I use a RXi-5ms 30meter, .053 mm ID with a guard column.
35C for 2 minutes rate 30degrees/min, final temp 320, final time 1 min. inj 250C and fid@320c I inject 1uL.
Hi everyone I have a strange issue to report. We are using a 6890GC with FID and TCD that was upgraded many years ago to give use LAN communication rather than GPIB. It works great even though it is over 20 years old. Just prior to Christmas we installed…
Hello from Kazakhstan,
Very interesting theme.
I also have some issues with flow method correction and some questions.
1) We use Excel CFT calculator. So how much pressure we should put on calculator, when using SCD detector with FID? As calculator…
This Information Applies To: Agilent ICP Expert software.
Issue
The software version is often asked and is a valuable information: to see if any updates are needed, if certain options are available etc.
Steps to Follow
Launch ICP Expert...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to calculate peak-to-valley ratio in OpenLab CDS.
Background
The peak-to-valley ratio is calculated to indicate the quality of peak separation...
This Information Applies To: Agilent 6500 Series Q-TOF LC/MS systems, MassHunter Workstation 10.1 or below and MassHunter Workstation 11.0 .
Issue
Fully reset the Q-TOF LC/MS system via a power cycle without venting. For users who experience slow...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II systems
Issue
The volume of sample used by Agilent parallel capillary electrophoresis systems during a run.
Resolution
The...
This Information Applies To: Agilent 2100 Expert software, 2100 Expert Security Pack, 2100 Bioanalyzer
Issue
When opening the 2100 Expert software and using the Security Pack, the error message " Start of launcher application failed - The requested...
This Information Applies To: Agilent 6890, 6850, 7890, 8890, 8860, Intuvo 9000 GC systems.
Issue
The exchange of the ignitor in a Flame Photometric Detector (FPD) may be required in the following scenarios:
Preventive or periodic maintenance...
This Information Applies To: Agilent 1290 Infinity Quaternary Pump G4204A, 1290 Infinity II Flexible Pumps G7104A/C, and G7131A/C.
Issue
If the flexible/quaternary pump does not generate flow or build up pressure, it could be due to a stuck inlet...
This Information Applies To: Agilent 5200, 5300 and 5400 Fragment Analyzer systems
Issue
Separation times for the Fragment Analyzer for each array format and kit.
Resolution
Caution : The times in this article are the length of time...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, Oligo Pro II, ProteoAnalyzer systems
Issue
Guidance for installing the Agilent parallel capillary electrophoresis instrument controller and/or...
This Information Applies To: OpenLab CDS (version 2.x) system
Issue:
This article explains how to export OpenLab CDS 2.x data as Archived file.
For copying OpenLab CDS 2.x data to another Projects or another OpenLab CDS systems, you can use data...