This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems
Issue
Autotune fails to complete.
Background
Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument. Sometimes when running the autotune procedure the autotune does not complete and the user is given an Autotune failed message in MassHunter. One of the main reasons for Autotune failure is a lack of sensitivity or abundance of the tuning ions. This article goes through some simple checks that a user can do to try to get the instrument to pass Autotune.
Items Required
Resolution / Steps to Follow
Figure 1 - Autotune window showing Autotune progress in the Red Box.
Figure 2 PEEL fitting, shown indicated by the red box, to check the drip rate of calibrant in step 2
Figure 3 - To start an Autotune from factory default parameters, check the circled Start from factory defaults box and then click the Autotune button.
Figure 4 - Screenshot of the Manual Tune tab showing the MS1 subtab settings (shown in red section).