This information applies to: Agilent LC and LC/MS instruments which use PEEK tubing.
Issue
Clean, straight edge cuts of PEEK tubing are essential to avoid chromatography and void volume issues.
Items required
PEEK tubing cutter (p/n 8710...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up a product ion acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Product ion...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 and later.
Issue
Set up a Neutral Gain/Loss acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Neutral...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 and later.
Issue
Procedure to create an acquisition method to submit injections.
Requirements
The instrument is configured and communicates with the computer...
This I
information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up an MRM acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select MRM, dMRM, or...