This Information Applies To: Agilent 6200 Series TOF and 6500 Series Q-TOF Systems
Issue
Instructions on how to correctly dilute the tune mix for 6200 Series TOF and 6500 Series Q-TOF systems depending on source type.
Background
If the Tune...
This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems
Issue
Autotune fails to complete.
Background
Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument...
This Information Applies To: Agilent 6200/6400/6500 Series LC/MS Systems
Issue
Instructions on how to fill the calibrant bottle on 6200/6400/6500 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal.
Issue:
Installing the Ion Transfer Capillary after cleaning or replacement. For instructions for installing an ion capillary...
This Information Applies To: Agilent 6000 Series LC/MS Systems equipped with MS40+ rotary vane pumps
Issue
Replacing the MS-40+ (rough pump) fluid periodically is a key maintenance procedure to improve vacuum performance and minimize the possibilities...