This Information Applies To: Agilent 6400 Triple Quadrupole LC/MS systems, MassHunter Workstation 10.1 or earlier. This article does not apply to MassHunter Workstation 12.0.
Issue
Users need to reconfigure a new LC/MS triple quadrupole after a new MassHunter software installation.
Reconfiguring the LC/MS triple quadrupole may be useful when troubleshooting the following software-related issues:
Steps to Follow
|
Caution: Do not skip steps 2 and 3 and go straight to step 4. Clicking Remove MassHunter Processes alone might cause corruption of tune files and/or methods.
|
Figure 1: The tool bar displaying System Engine menu, Shutdown Engines is the second option down
Figure 2: Run Remove MassHunter Processes
1. Agilent MassHunter Workstation folder, 2. Acq Tools folder, 3. MassHunter Processes
Figure 3: Instrument Configuration in version B.09.00 and 10.0 or above. Red x icons are highlighted with red box.
Figure 4: Instrument Configuration in version B.08.02 or below
Learn how to effectively operate your Agilent 6400 Series triple quadrupole LC/MS system:
Agilent 6400 Series Triple Quadrupole LC/MS with MassHunter Workstation e-learning path available from Agilent education |