This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up a product ion acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Product ion ( Figure 1 ).
Figure 1 . Scan type drop-down...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 and later.
Issue
Set up a Neutral Gain/Loss acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Neutral Gain or Neutral Loss ( Figure 1 ).
Figure...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 and later.
Issue
Procedure to create an acquisition method to submit injections.
Requirements
The instrument is configured and communicates with the computer and software.
Steps to Follow
Launch MassHunter...
This I
information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up an MRM acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select MRM, dMRM, or tMRM ( Figure 1 ).
Figure 1 . Scan Type drop-down...
This information applies to: Agilent Triple Quadrupole LC/MS System, MassHunter 12.0 or later.
Issue
Set up an MS2 scan acquisition method.
Steps to Follow
From the Scan type drop-down list in the Time Segments window, select Scan ( Figure 1 ).
Figure 1 . Scan Type drop-down list
...