This Information Applies To: MassHunter GC/MS Acquisition Versions B.07.xx through version 10.x.
Issue
MassHunter sequences stopping without errors, but “index out of bounds” is displayed at the bottom left of the acquisition pane.
Background
During acquisition, the sequence stopped with no errors in the logbook, no communications errors, and no hardware errors. The software displays the error index out of bounds at the bottom of the acquisition screen.
Figure 1. MassHunter Acquisition software displaying "index out of bounds" error in the lower left corner of the window during sequence run.
The use of a network printer is typically the cause of this error. If the print buffer is full, it will not respond in time and MassHunter will report this message.
Steps to Follow
Figure 2. Microsoft Print to PDF selected as the default printer.
Tip: Setting a Network printer as the default printer is not supported for MassHunter.
|