Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC/MS Portal
  • Agilent Knowledge Portal
LC/MS Portal
Articles How to Pump Down your Single Quadrupole LC/MS System Using OpenLab CDS
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC/MS Portal Home
  • LC/MS Portal Articles
  • LC/MS Portal Videos
  • More
  • Cancel
  • Created :  27 Feb 2023
  • Modified :  28 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  LC/MS Systems
  • Component :  LC/MS Single Quadrupole
  • Product Name :  6110 Single Quadrupole LC/MS 6120 Single Quadrupole LC/MS 6130 Single Quadrupole LC/MS 6140 Single Quadrupole LC/MS 6150 Single Quadrupole LC/MS InfinityLab LC/MSD InfinityLab LC/MSD XT InfinityLab LC/MSD iQ
  • Task :  Operation
Related Articles
  • Removing the Ion Transfer Capillary from Agilent 6000 Series LC/MS Systems

    This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here Issue: Removing the Ion Transfer Capillary for maintenance. Steps to follow: Tips : The figures in this article show an LC/MS with an Agilent Jet Stream (AJS) source installed. The process of capillary removal is the same for LC/MS instruments...
  • MS40-S and MS-40+ Foreline Pump Oil Change for ICP-MS Systems

    This Information Applies To: Agilent ICP-MS (7800/7850/7900/8900) Issue How to replace the Foreline Pump Oil in the MS-40+ or MS-40-S? Background The oil should be replaced at least once every - months, or when the oil color is indicates it needs to be changed . The foreline pump should be checked monthly. Items Required AVF Platinum Foreline Pump Oil (p/n X3760-64004 ) Oil Collection Container (at least 1 liter in size) Oil nozzle cap (or funnel) Paper towels 10 mm hex...
  • Filter replacement in the Agilent MS40+ Rotary Vane Pump

    This Information Applies To: Agilent 6000 Series LC/MS systems equipped with MS40+ rotary vane pumps Issue Instructions on how to replace the filter in an Agilent MS 40+ pump as part of annual pump maintenance Background The filter in the MS40+ pump should be replaced every 12 months of operation as part of an annual service on the pump. The pump fluid should also be replaced at this time (at least every 6 months), see Replacing the Pump Fluid in the MS-40+ Rotary Vane Pump detailed instructions...
  • How to Clean the AriaMx Cartridge Modules Carrier Rails

    This Information Applies To: Agilent AriaMx real-time PCR system Issue Agilent technical support may recommend cleaning the AriaMx's Cartridge Modules Carrier (OSM) rails as part of the troubleshooting process. Background The presence of dust and old grease on the AriaMx's OSM rails can have a negative impact on the instrument's performance. This debris can affect the movement of the OSM, causing problems such as suboptimal fluorescence patterns (e.g., fluctuations in the baseline signal...
  • How To Pump Down (Start) the GC/MSD - EI - MassHunter

    This Information Applies To: Agilent 5977A/B GC/MSD with MassHunter GC/MS Acquisition Software Issue: This article describes the procedure to safely pump down (start) 5977A/B GC/MSD with Electron Ionization (EI) Steps To follow: Before you turn on the MSD: The vent valve should be open slightly. All other vacuum seals and fittings must be in place and fastened correctly. The front side plate screw should not be tightened, unless hazardous carrier, JetClean system hydrogen, or reagent...

How to Pump Down your Single Quadrupole LC/MS System Using OpenLab CDS

Answer

This Information Applies To: 6100 Series single quadrupole LC/MS systems


Issue

Instructions for starting and pumping down your Agilent Single Quad 6100 series LC/MS instrument using OpenLab CDS software.


Requirements

This procedure starts with the MS vented and powered off.

Steps to Follow

  1. Make sure that the MS power cord is plugged into a grounded electrical outlet.

  2. Turn on the power to the instrument using the power switch on the bottom-left corner of the instrument (see highlighted section in Figure 1).

  3. Wait for a couple of minutes for the smartcard in the MS to start before communication between the MS and OpenLab is possible. You should hear the Calibrant Delivery System (CDS) valve switch position approximately one to three minutes after turning on the MS. This sound indicates that the MS smartcard has initialized and the MS is ready for communication and OpenLab control.


     Figure 1
    Figure 1. Power button location (bottom-left).

  4. Turn on each LC module while waiting for the MS Smartcard to initialize.

  5. Start the OpenLab Acquisition program.

  6. If not already visible, in the Home tab f the ribbon, click Instrument Status. (See Figure 2)


     Figure 2
    Figure 2. Instrument Status button location.

  7. If necessary, click Take to acquire control of the instrument. In Figure 3, Take is already selected.

     Figure 3
    Figure 3. Take instrument.

  8. In the Instrument Status window, click the Dashboard tab.

  9. Right-click the SQ device panel and click Pump Down.

      
    Caution: Tuning or trying to start analysis on an MS that is not sufficiently pumped down and equilibrated to temperature will result poor system performance, and the potential for instrument contamination and damage to the ion optics electronics. Do not rush the next step.
  10. Wait for your MS to complete the pump down process and reach your normal operating pressures.

    1. The rough vacuum should be between 1.0 - 2.5 Torr with a 0.5 mm I.D Ion capillary installed, or between 2.0 - 3.5 Torr with 0.6 mm I.D Ion capillary installed.

    2. The high vacuum should be 3.0 x 10-6 - 2.0 x 10-5 Torr

    3. Turbo speed from 97 to 103%

  11. Once the MS is sufficiently pumped down, load your required analytical method, and wait for your temperature and gas flow setpoints to be reached and heated zones to equilibrate.

      
    Caution: Wait at least 2 hours for all MS heated zones to fully equilibrate after reaching temperature setpoints before tuning your instrument to ensure a stable tune result.
  12. When ready for analysis, the SQ panel in the Instrument Status will turn green.

     Figure 4 
    Figure 4. SQ device status Ready.

  13. It is recommended after pumping down and allowing the MS to equilibrate, to run a tune on the instrument to check system performance before analyzing samples. If you have any issue with the pump down and start of your LC/MS instrument, then please contact your local Agilent Service representative.

 

  Learn how to effectively operate your Agilent Single Quadrupole LC/MS:
Agilent Single Quadrupole LC/MSD with OpenLab CDS e-learning path available from Agilent education