This Information Applies To: Agilent high performance autosamplers, and well plate samplers G1368A, G1377A, G1367A/B/C/D/E, G4226A, and G5667A
Issue
The bottom sensing feature in Agilent OpenLab Software is used to detect depths of vials or plates. Material of plates or vials must be able to resist the force of the needle; not all vials are suited.
Requirements
Compatible consumables for bottom sensing feature (Figure 1):
Incompatible consumables for bottom sensing feature (Figure 2):
Steps to Follow
The Vial/Well Bottom Sensing feature allows the needle to detect nonuniform well bottoms and adjusts the depth of the needle position above the detected bottom of the vial or well. You can use this feature in combination with the Draw Position to customize the draw position of the needle.
Right click the sampler, select Method, under Advanced > Auxiliary, check the box Vial/Well bottom sensing to turn on this feature (see Figure 3). The default setting is cleared:
Draw Position
The Draw Position is a vertical offset that enables you to position the needle a specific distance (in mm) away from its standard position. The Draw Position offset is calculated from the bottom of the well, with a slight amount of clearance to cover tolerances.
Table 1 shows needle tip position when draw position/ offset = 0
|
54 x 2 mL vials (tray) |
No bottom sensing |
4 mm |
With bottom sensing |
1 mm |
Tips: You may want to turn off Vial/Well Bottom Sensing to increase speed of injection, or to avoid the needle touching the bottom of the well if a sample precipitate could clog it.
|
Learn how to effectively operate your Agilent samplers: 1290 High Performance Autosampler InfinityLab LC Series user manual on Agilent.com 1260 Bio-inert High Performance Autosampler InfinityLab LC Series user manual on Agilent.com |