This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However, the original RFID tag is no longer available...
This Information Applies To: Agilent G4208A Instant Pilot
Issue
This article describes how to switch the illumination of the autosampler on or off (if available) using the G4208A Instant Pilot.
Items Required
G4208A Instant Pilot for Agilent LC
Steps to Follow
From the Welcome screen...
This Information Applies To: Agilent 1290 Infinity and 1290 Infinity II LC Pumps.
Issue
Solvent bottle runs dry. As the result, the pump has continued to run dry or solvent lines are filled with air. The difference between purge, condition, and prime functions and when to use them are explained...
This Information Applies To: Agilent LC Pumps with built-in Degasser
Issue
If the built-in degasser is not working and gives an error on power-on, the pump will be in error mode, and therefore no injections can be made. The user may wish to disable [or enable] the degasser, especially in an isocratic...
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the column in an appropriate solvent (check column information...