This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer
Issue:
Dirt or dust on the capillary array window can cause baseline shifts or elevation of the baseline. Figure 1 shows an example electropherogram with baseline shifts. It is sometimes possible to see the dirt or dust in the View Capillary Positions screen in the Agilent ProSize data analysis software. This screen can be opened by clicking Analysis>View Capillary Positions in ProSize. An example of dust on a capillary array is shown in Figure 2.
Figure 1. A portion of a Fragment Analyzer electropherogram with Baseline shifts indicated with blue arrows.
Figure 2. Fragment Analyzer capillary positions from the View Capillary Positions window in ProSize. The capillaries are depicted as distinct blue peaks. Dust will show as additional signal, see right-most peak.
If the baseline shift or elevation has occurred in just one run, the instrument ventilation between runs may be sufficient to remove the particles. However, if the baseline shifts or baseline elevation persists for multiple runs cleaning the capillary array window can be a useful troubleshooting step.
Steps to Follow:
Caution: Be gentle to avoid damage to the capillaries or capillary array.
|
Caution: It is important to let the capillaries air dry before re-attaching the light guide. The light guide can evaporate the alcohol solution resulting in alcohol condensing on the glass filter behind the array window.
|
If you had to adjust the capillary alignment for this separation, then it will be necessary to correct the alignment using the instrument controller software. Capillary alignment in the controller software will avoid the alignment being incorrect in future data files. Please check Chapter 4 of the Fragment Analyzer System Manual, Femto Pulse User Manual or ZAG DNA Analyzer System Manual for instructions. It is recommended and easiest to follow the Capillary Alignment from a File method. The capillary alignment should be done with the new data file, as only this latest run will have the current capillary positions.
Figure 3. The two white screws that secure the light guide to the window being removed on a Fragment Analyzer
Figure 4. Cleaning a Fragment Analyzer capillary array window
Figure 5. Fragment Analyzer or ZAG DNA Analyzer capillary array window with paper towel behind
Figure 6. Femto Pulse capillary array window with paper towel behind
Learn how to effectively operate your Agilent parallel capillary electrophoresis system:
|
For Research Use Only. Not for use in diagnostic procedures.
PR7000-8238