This Information Applies To: Agilent 1260 Infinity and 1260 Infinity II ELSD (G4260B)
Issue
When facing issues in the ELSD like broad peaks, or high pressure, it is important to identify its origin.
This information will help you to solve the issue, or to provide the Agilent Service Representative with the necessary information to do so.
Steps to Follow
ELSD high pressure
High-pressure issues are usually originated from a blocked nebulizer. A flush can remove the blockage.
Tips: Recommended solvents for the ELSD flush are Acetone/Water (50/50), Hexane/IPA (50/50), THF, DCM.
|
Evaporator temperature 40 °C | Nebulizer temperature 40 °C | Gas flow 1.6 SLM |
Table 1: Specified conditions for ELSD settings.
Increased noise/loss of signal
High background
It usually happens when the evaporator tube becomes contaminated with nonvolatile material resulting in high background or poor chromatography.
Tips: It is recommended to clean the evaporator tube once a week or every 40 hours of use as a preventive routine.
|
Poor RSD
Broad peaks
Reset gas to optics and nebulizer, then run the instrument until "Ready" status is achieved. Then check the autozero function.
In the previous recommendations do not help to fix the issue, contact your local Agilent support representative.