This Information Applies To: Agilent Intuvo 9000 GC and Agilent 5977A/B GC/MSD with MassHunter Data Acquisition software for GC/MS
Issue
In the middle of a sequence run, Masshunter Acquisition displays an error message "The instrument cannot be run when it is in the following state: FAST_VENT"...
This Information Applies To: Agilent GC/MSD systems
Issue:
When performing MS tuning, there is error message "No Emission Current" and tune is aborted.
Steps to Follow
This error normally means that the selected filament is not working. Perform the following checks:
In the Manual Tune...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD systems.
Issue
During tuning, MassHunter prompted the error message: "MSD Fault- RFPA Difficulty", and tune action stopped.
Steps to Follow
Try to reconfigure the GC/MS. See article: How To Configure a New GC/MSD with the...
This Information Applies To: MassHunter acquisition software for GC/MS systems and MSD Productivity ChemStation for all Agilent GC/MS systems
Issue
Manual tuning lets you interactively set the MSD parameters, such as lens voltages and tuning masses, to values that meet the needs of your particular...
This Information Applies To: Agilent Single Quadrupole GC/MSD Systems
Issue
Instructions to set the source and quadrupole temperatures on your GC/MSD using OpenLab CDS software.
Steps to Follow
Setpoints for the MSD ion source and mass filter (quadrupole) temperatures are stored in the current...