This Information Applies To: Agilent GC/MSD systems
Issue
Instructions on how to vent the Single Quad GC/MSD instrument using OpenLab CDS software.
Background
If the GC/MSD is running and under vacuum, then the analyzer needs to be vented before opening for maintenance, or to shut down the...
This Information Applies To: Agilent GC/MSD Systems using the Agilent IDP-3 dry scroll pump
Issue
Scroll pump requires replacement two tip seals and one O-ring after approximately one year of use (9000 hours).
Requirements
The system must be vented before starting this procedure. See the...
This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures. Do not touch any parts of GC and MS until...
This Information Applies To: Agilent 7000A/B/C/D Triple Quadrupole GC/MS systems
Issue
When a triple quadrupole GC/MS system tune fails, the user can check for any abnormalities by setting the tune file to default tune file before checking for hardware contamination or other issues.
Steps to...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving by GC inlet, or column...