Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Articles Region Information not Included in Agilent 2100 Bioanalyzer System PDF Report
  • Articles
  • Files
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Automated Electrophoresis Portal Home
  • Automated Electrophoresis Portal Articles
  • Automated Electrophoresis Portal Files
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Automated Electrophoresis Software Automated Electrophoresis Systems
  • Component :  Bioanalyzer Systems
  • Product Name :  2100 Bioanalyzer Instrument 2100 Expert Software
  • Task :  Data Analysis Reporting Troubleshooting
Related Articles
  • Troubleshooting Bioanalyzer Communication

    This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software Issue If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in this article to re-establish a stable instrument connection. Steps to Follow Instrument cannot connect or error: Instrument connection timeout Check if license keys have been registered in the software. Go to Help > Registration > Add Licenses...
  • How to Set Up a Region in the 2100 Expert Bioanalyzer Software

    This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert Software Issue Samples with a broad signal such as NGS smears should be analyzed by setting up a region. Analyzing a region is also termed smear analysis. Steps to Follow DNA, Small RNA, DV200 RNA, and Plant RNA assays For these data files, smear analysis is automatically enabled. For these assays, to add a region to a single sample, you can add a region as described in the following instructions and shown in...
  • Counter Mismatch Error in Agilent 2100 Bioanalyzer

    This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software without Security Pack addon Issue The 2100 Expert software gives a counter mismatch error in the log book which may lead to data loss (blank lanes). Specifically, it occurs when the 2100 Expert software on the PC detects that there are data points missing in the information it is receiving from the Bioanalyzer during a run. This indicates an interruption in the communication between the instrument and the PC...
  • DV200 Analysis of FFPE RNA Samples with the 2100 Bioanalyzer System

    This Information Applies To: 2100 Bioanalyzer, 2100 Expert software Issue The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents the percentage of RNA fragments larger than 200 nucleotides relative to all RNA fragments. The DV 200 of FFPE RNA samples can be calculated within the Agilent 2100 Expert software. Steps to follow Software version B.02.10 or higher If...
  • "Failed to Generate Report" Error When Creating a PDF Report on the 2100 Bioanalyzer

    This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software version B.02.08 or later Issue When printing a report to PDF or HTML with the 2100 Expert software, the error "Failed to Generate Report" appears. Background This error can appear because certain margins are not compatible with the document printer ( Figure 1 ). Figure 1 . Failed to Generate Report error Steps to Follow Modify the printer margin settings in the 2100 Expert software Click File > Page Setup...

Region Information not Included in Agilent 2100 Bioanalyzer System PDF Report

Answer

This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software


Issue:

The PDF report generated in the Agilent 2100 Expert software via File > Print with Result Tables checked as per Figure 1 did not include the regional table. 

 

Figure 1. Print menu


Resolution: 

The most likely root cause is an incorrect number format setting. To change it, follow these steps:

  1. Navigate to the Windows region settings. For Windows 10, type intl.cpl into the Windows search (see Figure 2), or search for region settings and click Additional date, time & regional settings .
     
    Figure 2. Access region settings
  2. The Format should be set to English (United States), see Figure 3. If the Format needs to be changed, please note that it influences other software as well (for example Microsoft Office applications). 
      
    Figure 3. Region settings menu
  3. Next, click Additional settings... (see Figure 3) and verify that the marked parameters are set like in Figure 4:

     

    Figure 4. Additional settings

If the language settings did not fix the issue, check the Page Setup in the Expert software. Open the 2100 Expert software, go to File > Page Setup, and check if the setup looks like in Figure 5:

  

Figure 5. Page Setup in the 2100 Expert software

 

 Tip 
Learn how to effectively maintain and troubleshoot your Agilent 2100 Bioanalyzer system:
Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide

 

For Research Use Only. Not for use in diagnostic procedures.

PR7000-7974

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help