This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
The PDF report generated in the Agilent 2100 Expert software via File > Print with Result Tables checked as per Figure 1 did not include the regional table.
Figure 1. Print menu
Resolution:
The most likely root cause is an incorrect number format setting. To change it, follow these steps:
Figure 2. Access region settings |
Figure 3. Region settings menu |
Figure 4. Additional settings |
If the language settings did not fix the issue, check the Page Setup in the Expert software. Open the 2100 Expert software, go to File > Page Setup, and check if the setup looks like in Figure 5:
Figure 5. Page Setup in the 2100 Expert software
Learn how to effectively maintain and troubleshoot your Agilent 2100 Bioanalyzer system:
|
For Research Use Only. Not for use in diagnostic procedures.