This Information Applies To: Agilent 4200 TapeStation 4150 TapeStation, TapeStation controller software
Issue
The system verification can be run to check the TapeStation performance, for example, when you suspect an instrument hardware fault.
Items Required
4200 TapeStation
TapeStation...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in this article to re-establish a stable instrument connection...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue:
This article explains how to install the Agilent 2100 Expert software and connect a PC to an Agilent 2100 Bioanalyzer system.
Steps to follow:
The latest version of the Expert software, can be downloaded free of charge...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system run ends prematurely and the following error appears in the logbook: Run aborted on port x. Instrument error occurred on port x, Unusual high or low voltage or current was detected during...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The Agilent 2100 Bioanalyzer system electropherograms in the Agilent 2100 Expert software have a wavy, uneven baseline ( Figure 1 ).
Figure 1. Wavy, uneven baseline on Bioanalyzer electropherograms
The error is most often...