Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Spectroscopy
  • Technical Areas
Atomic Spectroscopy
Wiki ICP-MS Active Cleaning
  • Announcements
  • Forum
  • Videos
  • Wiki
  • More
  • Cancel
Actions
  • Share
  • More
  • Cancel
Table of Contents
  • +Collection of Atomic Spectroscopy Resources Home
  • Basic Identification of Gas Box Type in AAS instrument for New Customers
  • Explore Atomic Spectroscopy eLearning Courses Available from Agilent Education
  • ICP-MS Active Cleaning
  • ICPMSD Cone Conditioning Procedure
  • Instructions for Long-term shut down of ICP-MS instruments
  • Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up
  • Tips and Tricks for ICP-MS General Applications
  • ICPMSD SDA Compliance Software Frequently Asked Questions
Still Need Help?

Post your question in our User Forum or Contact Support.

ICP-MS Active Cleaning

Created by Carlos Vargas Carlos Vargas over 2 years ago | Last modified by kristen.sensabaugh kristen.sensabaugh 1 month ago

 

This Information Applies To: All Agilent ICP-MS Systems

 

Issue:

Some samples may contain "Sticky" elements that take a long time to leave the system. After cleaning all glassware in the Sample Introduction System as per the maintenance DVD, an extra Active Clean can be performed. 

 

Resolution:

Use appropriate PPE when working with strong acids

  • When “sticky” elements (Tl, Sb, Th, Ag…) are taking too long to rinse out, replace autosampler probe and all other tubing leading up to the nebulizer.

 

  • Option 1 to clean the system: with the plasma off, remove PTFE line from back of nebulizer and hang in a beaker. With all of the sample introduction lines hooked up, move the autosampler probe into 50 ml vials of each of the following solutions:
    • ~30% NH4OH (30 min with peri at 0.3 rps)
    • ~30% HCl (30 min with peri at 0.3 rps)
    • ~30% HNO3 (30 min with peri at 0.3 rps)
    • De-ionized (a few minutes at 0.3 rps)
  • Let each of these solutions flow through the whole sample flow path. You might have to change the Tygon tubing after doing this.

Make sure you empty the 30% NH4OH from beaker before going to acid soluitons

  • Option 2 to clean the system: Clean the probe rinse port by removal then soaking or actively pumping for 15 minutes for each step using the ALS peristaltic pump using the following solutions:
    • MeOH or IPA to remove greasiness
    • De-ionized Water
    • Hot 2% Citrinox Solution
    • De-ionized Water
    • ~30% NH4OH
    • ~30% HCl
    • ~30% HNO3
    • De-ionized Water

 

  • Additional Cleaning Solutions (used in the Semiconductor industry). Use one or both of the following solutions, either as a rinse or for periodic cleaning

 

    

 

 

Learn more on how to effectively operate your Agilent ICP-MS Systems:

ICPMS-0GEN-1010e - Agilent ICP-MS with MassHunter: An Introduction to ICP-MS     e-Learning course on Agilent University 

ICPMS-0GEN-1030e - Agilent ICP-MS with MassHunter: Analysis Considerations  e-Learning course on Agilent University

 

Keywords: Relative Standard Deviation, RSD, memory, contamination, ICPMS, sq, qqq, 7500, 7700, 7800, 7900, 8800, 8900

 

  • 7900
  • 7700
  • 7800
  • SQ
  • 7500
  • ICPMS
  • 8800
  • 8900
  • Agilent Support Insights
  • Share
  • History
  • More
  • Cancel
  • Sign in to reply
Parents
  • cosmerosales
    cosmerosales over 2 years ago

    Thank you for this tips.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Comment
  • cosmerosales
    cosmerosales over 2 years ago

    Thank you for this tips.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Children
No Data
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Privacy Statement
Terms of Use
Contact Us
Site Help