Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Spectroscopy
  • Technical Areas
Atomic Spectroscopy
Wiki Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up
  • Announcements
  • Forum
  • Atomic Absorption Articles
  • ICP-OES Articles
  • ICP-MS Articles
  • Videos
  • Wiki
  • More
  • Cancel
Actions
  • Share
  • More
  • Cancel
Table of Contents
  • +Collection of Atomic Spectroscopy Resources Home
  • Basic Identification of Gas Box Type in AAS instrument for New Customers
  • Explore Atomic Spectroscopy eLearning Courses Available from Agilent Education
  • ICP-MS Active Cleaning
  • ICPMSD Cone Conditioning Procedure
  • Instructions for Long-term shut down of ICP-MS instruments
  • Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up
  • Tips and Tricks for ICP-MS General Applications
  • ICPMSD SDA Compliance Software Frequently Asked Questions
Still Need Help?

Post your question in our User Forum or Contact Support.

Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up

Created by Tina Harville Tina Harville over 3 years ago | Last modified by kristen.sensabaugh kristen.sensabaugh 7 months ago

Shutting Down Vista/700 Series Instrument for a Lengthy Period of Time:

  1. Turn the peristaltic pump on, and pump all liquid from the sample lines, nebulizer and spray chamber. Ensure you pump all the liquid out of the sample introduction components.
  2. Turn off the heat exchanger/chiller (water supply), then wait a few minutes.
  3. Turn off argon supply.
  4. Close the software and shut the instrument PC down.
  5. Switch off the power to the instrument (front left power switch).

 

Starting Up Vista/700 Series Instruments after a Lengthy Period of Shut-down:

  1. Switch the instrument power on (front left switch).
  2. Open the software.
  3. Turn on the argon, then wait a few minutes. (When argon is on, the minimum purge of 0.7L/min will be in effect while software is open. If needing to analyze wavelengths less than 190nm, turn the polyboost purge on in the software.)
  4. Turn on the heat exchanger/chiller (water supply).

 

Notes: When you repower the unit on, the optics will be at room temperature.  Turn the unit power on and allow the optics to warm for 1 day, prior to starting an analysis.

  • ICPOES
  • shutdown procedure
  • ICPOES chiller
  • Share
  • History
  • More
  • Cancel
  • Sign in to reply
  • arjun
    arjun over 3 years ago

    very useful for customer during this COVID-19 Period.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • sruddell
    sruddell over 3 years ago

    This is very helpful. Just a little note on the 700 and Vista the polypurge flows at 0.7l/min even if the instrument is off and the software does not affect this. Unlike the 51 and 58 /59 where the software has to be on and connected to enable the flow. 

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Privacy Statement
Terms of Use
Contact Us
Site Help