Wiki Tips and Tricks for ICP-MS General Applications
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
Atomic Spectroscopy requires membership for participation - click to join
  • Wiki
  • Alignment of the Sample Probe - PSD 120
  • How to Resolve SpectrAA Software Issues due to Adobe Flash Failure
  • Basic Identification of Gas Box Type in AAS instrument for New Customers
  • Collection of AAS & MP-AES Resources
  • Collection of ICP-OES Resources
  • Collection of ICPMS Resources
  • ICP-MS Active Cleaning
  • ICP-OES: How to Clean a Blocked Nebulizer
  • ICP-OES:  Checking for a Blocked Nebulizer
  • ICPMSD Cone Conditioning Procedure
  • Instructions for Long-term shut down of ICP-MS instruments
  • Long-term shutdown of Vista/700 Series ICPOES instruments and Eventual Start-up
  • Maintaining the Axial Pre-Optic Window
  • Tips and Tricks for ICP-MS General Applications
  • Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
  • Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components
  • Troubleshooting and Maintenance of AA Systems: Part 2B- Flame Optimization
  • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
  • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Tips and Tricks for ICP-MS General Applications

Agilent Support Insights

 

This Information Applies To: Agilent ICP-MS Systems

 

Issue:

This is a small collection of general application tips and tricks for ICP-MS

 

Recommendations:

Use appropriate PPE when working with strong acids

HCl and Polyatomic Interferences

Traditionally, HNO3 alone was preferred to avoid polyatomic interferences in ICP-MS without CRC. However, the lack of Cl- ion caused numerous problems with several elements. Therefore, we recommend using HCl in CRC system when you have:

  • Poor extraction efficiency during preparation (e.g., Sn in soil extracts).
  • Poor stability for many elements (Hg, Sn, Mo, W, Ag, As, Se, PGMs, REEs).
  • Poor stability of many elements in standard solutions due to lack of co-existing ions/ligands.
  • Poor washout characteristics.

 

Where possible, adding some HCl to the samples (0.5% minimum) may help with these cases.

 

Sample Container:

  • Where possible, use PP or PE (low density) for containers, avoid glass.
  • Dispose after use.
  • Polypropylene is better for long-term storage as it is less permeable.
  • Fluorinated polymers are an excellent choice as they are very clean.

 

HF use:

1% HF is sufficient to damage the standard sample introduction system

  • Please use the PFA kit (PFA spray chamber, PFA nebulizer, platinum cones).

 

 

Washout solutions:

  • Rinse port (HNO3)
  • Basic rinse (H2O, H2O2, NH4OH, EDTA, Triton X) stock solution
  • Acid mix (10% HCl, HNO3, 0.01% HF)
  • Acid matrix in sample

 

Solving problems with Ag and Pb analyses:

  • Though solubility rules say that some metals (Ag, Pb, Hg) precipitate as chlorides, due to presence of chloride ions in water. In water, both Ag and Pb would indeed precipitate as chlorides. The situation is different in acid. AgNO3 is water soluble, but Ag+ is easily reduced to Ag(metal). The reduction is acid catalyzed or photo catalyzed, so nitric acid alone is not good for Ag.
  • Though adding a little HCl may precipitate Ag+ as chloride, adding more brings it back into solution as [AgCl2]- or higher Cl- complexes, depending on the Cl- concentration. By having Ag complexed with HCl as a source of Cl-, you shift the equilibrium away from free Ag+ ion to silver chloro-complexes. This slows the rate of silver loss by reduction.
  • Pb is less of a problem. It is fine in diluted nitric acid. Since you will have Pb in acid, the water solubility rules again do not apply unless Pb and Ag concentrations are way above solubility levels anyway, typically not the case for analytical work. At any HCl concentration above 0.005 M, Pb2+ forms a soluble anionic complex.
  • Something that might work even better than HCl in a final dilution is getting your nitric acid down to 1% or lower with 0.1% cysteine. This acid-soluble sulfur ligand will grab those soft acid cation species (Pb and Ag) and hold them in solution. The problem is that higher nitric acid concentrations will oxidize the cysteine at a rate dependent on nitric acid concentrations. Immediately after dissolution, cysteine will be oxidized to cystine (the disulfide bonded dimer). Prolonged solution in strong oxidizer like nitric acid will take the oxidation further, diminishing its chelating power. That is why less than 1% nitric acid is good when using cysteine. So, it is not recommended to add cysteine before final dilution, or to the undiluted digest.

Learn more on how to effectively operate your Agilent ICP-MS Systems:

ICPMS-0GEN-1010e - Agilent ICP-MS with MassHunter: An Introduction to ICP-MS e-Learning course on Agilent University 

ICPMS-0GEN-1030e - Agilent ICP-MS with MassHunter: Analysis Considerations    e-Learning course on Agilent University

 

Keywords: ICPMS, SQ, QQQ, 7500, 7700, 7800, 7900, 8800, 8900

 

 

  • 7900
  • 7700
  • qqq
  • 7800
  • ICP-MS
  • sq
  • 7500
  • ICPMS
  • 8800
  • 8900
  • agilent support insights
  • Share
  • History
  • More
  • Cancel
Anonymous

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Related
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help